referenced SOP instance UIDs for multi-frame segmentations
Moderator: Moderator Team
referenced SOP instance UIDs for multi-frame segmentations
I am trying to read multi-frame Dicom segmentation objects (DSOs) using DCMTK. My DSO file stores several frames, each one annotating a different DICOM file from the series. I can see how to use the DcmSegmentation class to get basic image data, but I am stuck trying to read the metadata. My DSO file does not define the Referenced SOP Instance UID (0008,1155) tag. Instead there is some sort of tree, where each frame has separate metadata. I have tried calling DcmSegmentation::getReferencedPPS(), but it returns empty strings. Does anyone have advice for getting the referenced UIDs for each frame?
-
- DCMTK Developer
- Posts: 2051
- Joined: Fri, 2004-11-05, 13:47
- Location: Oldenburg, Germany
- Contact:
Re: referenced SOP instance UIDs for multi-frame segmentatio
Hi,
which attributes are the references in, i.e. the tag names and if applicable the sequence(s) where they are located.
If you are allowed to provide a copy of the file, you can send it to onken at open-connections de.
Thank you,
Michael
which attributes are the references in, i.e. the tag names and if applicable the sequence(s) where they are located.
If you are allowed to provide a copy of the file, you can send it to onken at open-connections de.
Thank you,
Michael
Re: referenced SOP instance UIDs for multi-frame segmentatio
Thanks for your help. I am trying to read the ReferencedSOPInstanceUID tag (0008,1155). This is stored in an Item (fffe,e000), which is in turn stored in a ReferencedInstanceSequence (0008,114a), which is itself stored in a ReferencedSeriesSequence (0008,1115). I have sent this file to oken@open-connections.de. Here is the relevant portion of the dcmdump output:
(0008,1115) SQ (Sequence with undefined length #=1) # u/l, 1 ReferencedSeriesSequence
(fffe,e000) na (Item with undefined length #=2) # u/l, 1 Item
(0008,114a) SQ (Sequence with undefined length #=22) # u/l, 1 ReferencedInstanceSequence
(fffe,e000) na (Item with undefined length #=2) # u/l, 1 Item
(0008,1150) UI =CTImageStorage # 26, 1 ReferencedSOPClassUID
(0008,1155) UI [1.2.826.0.1.3680043.8.420.18008083694064501353168288758878365437] # 64, 1 ReferencedSOPInstanceUID
(fffe,e00d) na (ItemDelimitationItem) # 0, 0 ItemDelimitationItem
(fffe,e000) na (Item with undefined length #=2) # u/l, 1 Item
(0008,1150) UI =CTImageStorage # 26, 1 ReferencedSOPClassUID
(0008,1155) UI [1.2.826.0.1.3680043.8.420.34854185099322246661873681910480156760] # 64, 1 ReferencedSOPInstanceUID
(fffe,e00d) na (ItemDelimitationItem) # 0, 0 ItemDelimitationItem
(fffe,e000) na (Item with undefined length #=2) # u/l, 1 Item
(0008,1150) UI =CTImageStorage # 26, 1 ReferencedSOPClassUID
(0008,1155) UI [1.2.826.0.1.3680043.8.420.42745719933517885967150852506372484375] # 64, 1 ReferencedSOPInstanceUID
(fffe,e00d) na (ItemDelimitationItem) # 0, 0 ItemDelimitationItem
(fffe,e000) na (Item with undefined length #=2) # u/l, 1 Item
(0008,1150) UI =CTImageStorage # 26, 1 ReferencedSOPClassUID
(0008,1155) UI [1.2.826.0.1.3680043.8.420.12830839117102264121920466256127611408] # 64, 1 ReferencedSOPInstanceUID
(fffe,e00d) na (ItemDelimitationItem) # 0, 0 ItemDelimitationItem
(fffe,e000) na (Item with undefined length #=2) # u/l, 1 Item
(0008,1150) UI =CTImageStorage # 26, 1 ReferencedSOPClassUID
(0008,1155) UI [1.2.826.0.1.3680043.8.420.22236842338853616830199403197384154638] # 64, 1 ReferencedSOPInstanceUID
(fffe,e00d) na (ItemDelimitationItem) # 0, 0 ItemDelimitationItem
(fffe,e000) na (Item with undefined length #=2) # u/l, 1 Item
(0008,1150) UI =CTImageStorage # 26, 1 ReferencedSOPClassUID
(0008,1155) UI [1.2.826.0.1.3680043.8.420.20854655349946513539015107098173610244] # 64, 1 ReferencedSOPInstanceUID
(fffe,e00d) na (ItemDelimitationItem) # 0, 0 ItemDelimitationItem
(fffe,e000) na (Item with undefined length #=2) # u/l, 1 Item
(0008,1150) UI =CTImageStorage # 26, 1 ReferencedSOPClassUID
(0008,1155) UI [1.2.826.0.1.3680043.8.420.20135966694136174446684606126183789649] # 64, 1 ReferencedSOPInstanceUID
(fffe,e00d) na (ItemDelimitationItem) # 0, 0 ItemDelimitationItem
(fffe,e000) na (Item with undefined length #=2) # u/l, 1 Item
(0008,1150) UI =CTImageStorage # 26, 1 ReferencedSOPClassUID
(0008,1155) UI [1.2.826.0.1.3680043.8.420.11839317830910667575807573412785071710] # 64, 1 ReferencedSOPInstanceUID
(fffe,e00d) na (ItemDelimitationItem) # 0, 0 ItemDelimitationItem
(fffe,e000) na (Item with undefined length #=2) # u/l, 1 Item
(0008,1150) UI =CTImageStorage # 26, 1 ReferencedSOPClassUID
(0008,1155) UI [1.2.826.0.1.3680043.8.420.28026964350116403778418776913022193556] # 64, 1 ReferencedSOPInstanceUID
(fffe,e00d) na (ItemDelimitationItem) # 0, 0 ItemDelimitationItem
(fffe,e000) na (Item with undefined length #=2) # u/l, 1 Item
(0008,1150) UI =CTImageStorage # 26, 1 ReferencedSOPClassUID
(0008,1155) UI [1.2.826.0.1.3680043.8.420.4926458493811260196158475672711346085] # 64, 1 ReferencedSOPInstanceUID
(fffe,e00d) na (ItemDelimitationItem) # 0, 0 ItemDelimitationItem
(fffe,e000) na (Item with undefined length #=2) # u/l, 1 Item
(0008,1150) UI =CTImageStorage # 26, 1 ReferencedSOPClassUID
(0008,1155) UI [1.2.826.0.1.3680043.8.420.11268513745288743682203776882717427048] # 64, 1 ReferencedSOPInstanceUID
(fffe,e00d) na (ItemDelimitationItem) # 0, 0 ItemDelimitationItem
(fffe,e000) na (Item with undefined length #=2) # u/l, 1 Item
(0008,1150) UI =CTImageStorage # 26, 1 ReferencedSOPClassUID
(0008,1155) UI [1.2.826.0.1.3680043.8.420.14071385335516728770730047221013460626] # 64, 1 ReferencedSOPInstanceUID
(fffe,e00d) na (ItemDelimitationItem) # 0, 0 ItemDelimitationItem
(fffe,e000) na (Item with undefined length #=2) # u/l, 1 Item
(0008,1150) UI =CTImageStorage # 26, 1 ReferencedSOPClassUID
(0008,1155) UI [1.2.826.0.1.3680043.8.420.18815256456037291279788197899925309626] # 64, 1 ReferencedSOPInstanceUID
(fffe,e00d) na (ItemDelimitationItem) # 0, 0 ItemDelimitationItem
(fffe,e000) na (Item with undefined length #=2) # u/l, 1 Item
(0008,1150) UI =CTImageStorage # 26, 1 ReferencedSOPClassUID
(0008,1155) UI [1.2.826.0.1.3680043.8.420.27992562356396331489017548480748247639] # 64, 1 ReferencedSOPInstanceUID
(fffe,e00d) na (ItemDelimitationItem) # 0, 0 ItemDelimitationItem
(fffe,e000) na (Item with undefined length #=2) # u/l, 1 Item
(0008,1150) UI =CTImageStorage # 26, 1 ReferencedSOPClassUID
(0008,1155) UI [1.2.826.0.1.3680043.8.420.24748395979620986642074560996190829556] # 64, 1 ReferencedSOPInstanceUID
(fffe,e00d) na (ItemDelimitationItem) # 0, 0 ItemDelimitationItem
(fffe,e000) na (Item with undefined length #=2) # u/l, 1 Item
(0008,1150) UI =CTImageStorage # 26, 1 ReferencedSOPClassUID
(0008,1155) UI [1.2.826.0.1.3680043.8.420.27504854622668372355151549907721499603] # 64, 1 ReferencedSOPInstanceUID
(fffe,e00d) na (ItemDelimitationItem) # 0, 0 ItemDelimitationItem
(fffe,e000) na (Item with undefined length #=2) # u/l, 1 Item
(0008,1150) UI =CTImageStorage # 26, 1 ReferencedSOPClassUID
(0008,1155) UI [1.2.826.0.1.3680043.8.420.61434704911503588225855376482709989624] # 64, 1 ReferencedSOPInstanceUID
(fffe,e00d) na (ItemDelimitationItem) # 0, 0 ItemDelimitationItem
(fffe,e000) na (Item with undefined length #=2) # u/l, 1 Item
(0008,1150) UI =CTImageStorage # 26, 1 ReferencedSOPClassUID
(0008,1155) UI [1.2.826.0.1.3680043.8.420.12959654711730148422891724111411962495] # 64, 1 ReferencedSOPInstanceUID
(fffe,e00d) na (ItemDelimitationItem) # 0, 0 ItemDelimitationItem
(fffe,e000) na (Item with undefined length #=2) # u/l, 1 Item
(0008,1150) UI =CTImageStorage # 26, 1 ReferencedSOPClassUID
(0008,1155) UI [1.2.826.0.1.3680043.8.420.23801707800765878351578521027518958052] # 64, 1 ReferencedSOPInstanceUID
(fffe,e00d) na (ItemDelimitationItem) # 0, 0 ItemDelimitationItem
(fffe,e000) na (Item with undefined length #=2) # u/l, 1 Item
(0008,1150) UI =CTImageStorage # 26, 1 ReferencedSOPClassUID
(0008,1155) UI [1.2.826.0.1.3680043.8.420.29957302060522776411053373218204619774] # 64, 1 ReferencedSOPInstanceUID
(fffe,e00d) na (ItemDelimitationItem) # 0, 0 ItemDelimitationItem
(fffe,e000) na (Item with undefined length #=2) # u/l, 1 Item
(0008,1150) UI =CTImageStorage # 26, 1 ReferencedSOPClassUID
(0008,1155) UI [1.2.826.0.1.3680043.8.420.21040073326579522610440371221121586421] # 64, 1 ReferencedSOPInstanceUID
(fffe,e00d) na (ItemDelimitationItem) # 0, 0 ItemDelimitationItem
(fffe,e000) na (Item with undefined length #=2) # u/l, 1 Item
(0008,1150) UI =CTImageStorage # 26, 1 ReferencedSOPClassUID
(0008,1155) UI [1.2.826.0.1.3680043.8.420.10300849820926352148110389939870185526] # 64, 1 ReferencedSOPInstanceUID
(fffe,e00d) na (ItemDelimitationItem) # 0, 0 ItemDelimitationItem
(fffe,e0dd) na (SequenceDelimitationItem) # 0, 0 SequenceDelimitationItem
(0020,000e) UI [1.2.826.0.1.3680043.8.420.61681108914021893915887608173970604170] # 64, 1 SeriesInstanceUID
(fffe,e00d) na (ItemDelimitationItem) # 0, 0 ItemDelimitationItem
(fffe,e0dd) na (SequenceDelimitationItem) # 0, 0 SequenceDelimitationItem
(0008,1115) SQ (Sequence with undefined length #=1) # u/l, 1 ReferencedSeriesSequence
(fffe,e000) na (Item with undefined length #=2) # u/l, 1 Item
(0008,114a) SQ (Sequence with undefined length #=22) # u/l, 1 ReferencedInstanceSequence
(fffe,e000) na (Item with undefined length #=2) # u/l, 1 Item
(0008,1150) UI =CTImageStorage # 26, 1 ReferencedSOPClassUID
(0008,1155) UI [1.2.826.0.1.3680043.8.420.18008083694064501353168288758878365437] # 64, 1 ReferencedSOPInstanceUID
(fffe,e00d) na (ItemDelimitationItem) # 0, 0 ItemDelimitationItem
(fffe,e000) na (Item with undefined length #=2) # u/l, 1 Item
(0008,1150) UI =CTImageStorage # 26, 1 ReferencedSOPClassUID
(0008,1155) UI [1.2.826.0.1.3680043.8.420.34854185099322246661873681910480156760] # 64, 1 ReferencedSOPInstanceUID
(fffe,e00d) na (ItemDelimitationItem) # 0, 0 ItemDelimitationItem
(fffe,e000) na (Item with undefined length #=2) # u/l, 1 Item
(0008,1150) UI =CTImageStorage # 26, 1 ReferencedSOPClassUID
(0008,1155) UI [1.2.826.0.1.3680043.8.420.42745719933517885967150852506372484375] # 64, 1 ReferencedSOPInstanceUID
(fffe,e00d) na (ItemDelimitationItem) # 0, 0 ItemDelimitationItem
(fffe,e000) na (Item with undefined length #=2) # u/l, 1 Item
(0008,1150) UI =CTImageStorage # 26, 1 ReferencedSOPClassUID
(0008,1155) UI [1.2.826.0.1.3680043.8.420.12830839117102264121920466256127611408] # 64, 1 ReferencedSOPInstanceUID
(fffe,e00d) na (ItemDelimitationItem) # 0, 0 ItemDelimitationItem
(fffe,e000) na (Item with undefined length #=2) # u/l, 1 Item
(0008,1150) UI =CTImageStorage # 26, 1 ReferencedSOPClassUID
(0008,1155) UI [1.2.826.0.1.3680043.8.420.22236842338853616830199403197384154638] # 64, 1 ReferencedSOPInstanceUID
(fffe,e00d) na (ItemDelimitationItem) # 0, 0 ItemDelimitationItem
(fffe,e000) na (Item with undefined length #=2) # u/l, 1 Item
(0008,1150) UI =CTImageStorage # 26, 1 ReferencedSOPClassUID
(0008,1155) UI [1.2.826.0.1.3680043.8.420.20854655349946513539015107098173610244] # 64, 1 ReferencedSOPInstanceUID
(fffe,e00d) na (ItemDelimitationItem) # 0, 0 ItemDelimitationItem
(fffe,e000) na (Item with undefined length #=2) # u/l, 1 Item
(0008,1150) UI =CTImageStorage # 26, 1 ReferencedSOPClassUID
(0008,1155) UI [1.2.826.0.1.3680043.8.420.20135966694136174446684606126183789649] # 64, 1 ReferencedSOPInstanceUID
(fffe,e00d) na (ItemDelimitationItem) # 0, 0 ItemDelimitationItem
(fffe,e000) na (Item with undefined length #=2) # u/l, 1 Item
(0008,1150) UI =CTImageStorage # 26, 1 ReferencedSOPClassUID
(0008,1155) UI [1.2.826.0.1.3680043.8.420.11839317830910667575807573412785071710] # 64, 1 ReferencedSOPInstanceUID
(fffe,e00d) na (ItemDelimitationItem) # 0, 0 ItemDelimitationItem
(fffe,e000) na (Item with undefined length #=2) # u/l, 1 Item
(0008,1150) UI =CTImageStorage # 26, 1 ReferencedSOPClassUID
(0008,1155) UI [1.2.826.0.1.3680043.8.420.28026964350116403778418776913022193556] # 64, 1 ReferencedSOPInstanceUID
(fffe,e00d) na (ItemDelimitationItem) # 0, 0 ItemDelimitationItem
(fffe,e000) na (Item with undefined length #=2) # u/l, 1 Item
(0008,1150) UI =CTImageStorage # 26, 1 ReferencedSOPClassUID
(0008,1155) UI [1.2.826.0.1.3680043.8.420.4926458493811260196158475672711346085] # 64, 1 ReferencedSOPInstanceUID
(fffe,e00d) na (ItemDelimitationItem) # 0, 0 ItemDelimitationItem
(fffe,e000) na (Item with undefined length #=2) # u/l, 1 Item
(0008,1150) UI =CTImageStorage # 26, 1 ReferencedSOPClassUID
(0008,1155) UI [1.2.826.0.1.3680043.8.420.11268513745288743682203776882717427048] # 64, 1 ReferencedSOPInstanceUID
(fffe,e00d) na (ItemDelimitationItem) # 0, 0 ItemDelimitationItem
(fffe,e000) na (Item with undefined length #=2) # u/l, 1 Item
(0008,1150) UI =CTImageStorage # 26, 1 ReferencedSOPClassUID
(0008,1155) UI [1.2.826.0.1.3680043.8.420.14071385335516728770730047221013460626] # 64, 1 ReferencedSOPInstanceUID
(fffe,e00d) na (ItemDelimitationItem) # 0, 0 ItemDelimitationItem
(fffe,e000) na (Item with undefined length #=2) # u/l, 1 Item
(0008,1150) UI =CTImageStorage # 26, 1 ReferencedSOPClassUID
(0008,1155) UI [1.2.826.0.1.3680043.8.420.18815256456037291279788197899925309626] # 64, 1 ReferencedSOPInstanceUID
(fffe,e00d) na (ItemDelimitationItem) # 0, 0 ItemDelimitationItem
(fffe,e000) na (Item with undefined length #=2) # u/l, 1 Item
(0008,1150) UI =CTImageStorage # 26, 1 ReferencedSOPClassUID
(0008,1155) UI [1.2.826.0.1.3680043.8.420.27992562356396331489017548480748247639] # 64, 1 ReferencedSOPInstanceUID
(fffe,e00d) na (ItemDelimitationItem) # 0, 0 ItemDelimitationItem
(fffe,e000) na (Item with undefined length #=2) # u/l, 1 Item
(0008,1150) UI =CTImageStorage # 26, 1 ReferencedSOPClassUID
(0008,1155) UI [1.2.826.0.1.3680043.8.420.24748395979620986642074560996190829556] # 64, 1 ReferencedSOPInstanceUID
(fffe,e00d) na (ItemDelimitationItem) # 0, 0 ItemDelimitationItem
(fffe,e000) na (Item with undefined length #=2) # u/l, 1 Item
(0008,1150) UI =CTImageStorage # 26, 1 ReferencedSOPClassUID
(0008,1155) UI [1.2.826.0.1.3680043.8.420.27504854622668372355151549907721499603] # 64, 1 ReferencedSOPInstanceUID
(fffe,e00d) na (ItemDelimitationItem) # 0, 0 ItemDelimitationItem
(fffe,e000) na (Item with undefined length #=2) # u/l, 1 Item
(0008,1150) UI =CTImageStorage # 26, 1 ReferencedSOPClassUID
(0008,1155) UI [1.2.826.0.1.3680043.8.420.61434704911503588225855376482709989624] # 64, 1 ReferencedSOPInstanceUID
(fffe,e00d) na (ItemDelimitationItem) # 0, 0 ItemDelimitationItem
(fffe,e000) na (Item with undefined length #=2) # u/l, 1 Item
(0008,1150) UI =CTImageStorage # 26, 1 ReferencedSOPClassUID
(0008,1155) UI [1.2.826.0.1.3680043.8.420.12959654711730148422891724111411962495] # 64, 1 ReferencedSOPInstanceUID
(fffe,e00d) na (ItemDelimitationItem) # 0, 0 ItemDelimitationItem
(fffe,e000) na (Item with undefined length #=2) # u/l, 1 Item
(0008,1150) UI =CTImageStorage # 26, 1 ReferencedSOPClassUID
(0008,1155) UI [1.2.826.0.1.3680043.8.420.23801707800765878351578521027518958052] # 64, 1 ReferencedSOPInstanceUID
(fffe,e00d) na (ItemDelimitationItem) # 0, 0 ItemDelimitationItem
(fffe,e000) na (Item with undefined length #=2) # u/l, 1 Item
(0008,1150) UI =CTImageStorage # 26, 1 ReferencedSOPClassUID
(0008,1155) UI [1.2.826.0.1.3680043.8.420.29957302060522776411053373218204619774] # 64, 1 ReferencedSOPInstanceUID
(fffe,e00d) na (ItemDelimitationItem) # 0, 0 ItemDelimitationItem
(fffe,e000) na (Item with undefined length #=2) # u/l, 1 Item
(0008,1150) UI =CTImageStorage # 26, 1 ReferencedSOPClassUID
(0008,1155) UI [1.2.826.0.1.3680043.8.420.21040073326579522610440371221121586421] # 64, 1 ReferencedSOPInstanceUID
(fffe,e00d) na (ItemDelimitationItem) # 0, 0 ItemDelimitationItem
(fffe,e000) na (Item with undefined length #=2) # u/l, 1 Item
(0008,1150) UI =CTImageStorage # 26, 1 ReferencedSOPClassUID
(0008,1155) UI [1.2.826.0.1.3680043.8.420.10300849820926352148110389939870185526] # 64, 1 ReferencedSOPInstanceUID
(fffe,e00d) na (ItemDelimitationItem) # 0, 0 ItemDelimitationItem
(fffe,e0dd) na (SequenceDelimitationItem) # 0, 0 SequenceDelimitationItem
(0020,000e) UI [1.2.826.0.1.3680043.8.420.61681108914021893915887608173970604170] # 64, 1 SeriesInstanceUID
(fffe,e00d) na (ItemDelimitationItem) # 0, 0 ItemDelimitationItem
(fffe,e0dd) na (SequenceDelimitationItem) # 0, 0 SequenceDelimitationItem
-
- DCMTK Developer
- Posts: 2051
- Joined: Fri, 2004-11-05, 13:47
- Location: Oldenburg, Germany
- Contact:
Re: referenced SOP instance UIDs for multi-frame segmentatio
Hi,
sorry for the late feedback: You can use DcmSegmentation's method getCommonInstanceReference() which DcmSegmentation inherits from DcmIODCommon. The API documentation of the returned class type IODCommonInstanceReferenceModule is also online.
Best,
Michael
sorry for the late feedback: You can use DcmSegmentation's method getCommonInstanceReference() which DcmSegmentation inherits from DcmIODCommon. The API documentation of the returned class type IODCommonInstanceReferenceModule is also online.
Best,
Michael
Re: referenced SOP instance UIDs for multi-frame segmentatio
Thanks, that worked perfectly!