i tried dcmdump +w command and it make me 54 files with .raw extension how to open or change transfer syntax from that?
-
- DCMTK Developer
- Posts: 2506
- Joined: Tue, 2011-05-03, 14:38
- Location: Oldenburg, Germany
- Contact:
Re: Dicom images cannot be read
You should change the Transfer Syntax UID in the file meta information header, i.e. in the text dump (redirect the console output to a text file for this purpose). The RAW files contain the pixel data. They do not need to be changed. They are just referenced from the dump and will be added to the output file when calling dump2dcm.
-
- Posts: 24
- Joined: Fri, 2016-12-02, 09:51
Re: Dicom images cannot be read
i tried to do that but i have some tags VR = ?? and when i tried to make dcm from dump i got this message:
VR of Tag (2001,1063) is unknown, using 'UN' for this data element
E: dump2dcm: dump.txt: Error in creating Element: Corrupted data (line 102)
W: VR of Tag (2001,10a1) is unknown, using 'UN' for this data element
E: dump2dcm: dump.txt: Error in creating Element: Corrupted data (line 103)
W: VR of Tag (2001,10c1) is unknown, using 'UN' for this data element
VR of Tag (2001,1063) is unknown, using 'UN' for this data element
E: dump2dcm: dump.txt: Error in creating Element: Corrupted data (line 102)
W: VR of Tag (2001,10a1) is unknown, using 'UN' for this data element
E: dump2dcm: dump.txt: Error in creating Element: Corrupted data (line 103)
W: VR of Tag (2001,10c1) is unknown, using 'UN' for this data element
-
- Posts: 24
- Joined: Fri, 2016-12-02, 09:51
Re: Dicom images cannot be read
Thanks for you help
I replaced all ?? with OB and i add transfer syntax jpeg lossless and work.
I replaced all ?? with OB and i add transfer syntax jpeg lossless and work.
-
- DCMTK Developer
- Posts: 2506
- Joined: Tue, 2011-05-03, 14:38
- Location: Oldenburg, Germany
- Contact:
Re: Dicom images cannot be read
If you replace " ?? " with " UN " in the dump file, it should work.
-
- DCMTK Developer
- Posts: 2506
- Joined: Tue, 2011-05-03, 14:38
- Location: Oldenburg, Germany
- Contact:
Re: Dicom images cannot be read
Great. I'll enhance "dump2dcm" right now, so that there is no need to replace the "??"