I have a file that used to be able to open with 3.5.4, but no longer opens with 3.6.0.

This is what dcmdump reports now (but doesn't report with 3.5.4):

W: DcmSequenceOfItems: Length of item in sequence StackSequence (2005,105f) is odd
E: DcmSequenceOfItems: Parse error in sequence, found (4e55,4e4b) instead of a sequence delimiter
E: dcmdump: Corrupted data: reading file: test.dcm

If 3.6.0 is so strict as to not open files that traditionally it did, I think I'll have to go back to 3.5.4, as it's difficult to explain to customers that they have to "fix" their scanners to accommodate new software...

Have the rules been really tightened with 3.6.0?


Top
 Profile  
 
 Post subject:
PostPosted: Sat, 2011-02-26, 20:31 
Offline
ICSMED DICOM Services
ICSMED DICOM Services

Joined: Fri, 2004-10-29, 21:38
Posts: 2217
Location: Oldenburg, Germany
I haven't checked the source code and also did not compare the version 3.5.4 implementation with the 3.6.0 one, but it seems that the sequence mentioned in the first message (level "WARN") has an odd length which is strictly forbidden in DICOM - this is also true for private elements as in your case.

If the corrupted file is accepted by DCMTK 3.5.4, are there any messages reported during parsing?


Top
 Profile  
 
 Post subject:
PostPosted: Sat, 2011-02-26, 21:30 
Offline

Joined: Tue, 2007-05-08, 17:45
Posts: 183
I understand that the file very likely breaks some DICOM rules, but my point was that something that worked fine with 3.5.4 no longer works.

I don't have the dcmdump report with me, I'll check on it Monday, but I don't recall seeing any messages (although I only took a quick glance).


Top
 Profile  
 
 Post subject:
PostPosted: Sat, 2011-02-26, 22:25 
Offline
ICSMED DICOM Services
ICSMED DICOM Services

Joined: Fri, 2004-10-29, 21:38
Posts: 2217
Location: Oldenburg, Germany
Please enable the verbose and debug mode with dcmdump, i.e. use options "-v" and "-d".


Top
 Profile  
 
 Post subject:
PostPosted: Mon, 2011-02-28, 09:28 
Offline

Joined: Mon, 2007-09-03, 10:53
Posts: 99
Location: Trondheim, Norway
Refusing to parse DICOM files with odd length elements is just plain silly. This arcane and utterly useless requirement should have been dumped from the standard ages ago. 16 bit alignment might have made sense in the 1980s, but today the padding does more harm than good, even on tiny embedded platforms.


Top
 Profile  
 
 Post subject:
PostPosted: Mon, 2011-02-28, 09:53 
Offline
ICSMED DICOM Services
ICSMED DICOM Services

Joined: Fri, 2004-10-29, 21:38
Posts: 2217
Location: Oldenburg, Germany
dcmdump and other DCMTK tools based on dcmdata (usually) do not reject DICOM files that contain odd length data elements, see option --accept-odd-length which is the default. That's the reason why we would be interested to compare the debug output of dcmdump 3.5.4 with the one from dcmdump 3.6.0 on this particular file.

Regarding your comment on removing this requirement from the DICOM standard: There is the "sacred cow" of backward-compatibility because of the large installed base ...


Top
 Profile  
 
 Post subject:
PostPosted: Mon, 2011-02-28, 16:50 
Offline

Joined: Tue, 2007-05-08, 17:45
Posts: 183
Below is the output from dcmdump 3.5.4 with the -d option (-v didn't exist).

Do you want the output from version 3.6.0 as well?


# Dicom-File-Format

# Dicom-Meta-Information-Header
# Used TransferSyntax: LittleEndianExplicit
(0002,0000) UL 198 # 4, 1 MetaElementGroupLength
(0002,0001) OB 01\00 # 2, 1 FileMetaInformationVersion
(0002,0002) UI =MRImageStorage # 26, 1 MediaStorageSOPClassUID
(0002,0003) UI [1.3.46.670589.11.17445.5.0.3368.2009110613452637003] # 52, 1 MediaStorageSOPInstanceUID
(0002,0010) UI =LittleEndianImplicit # 18, 1 TransferSyntaxUID
(0002,0012) UI [1.2.804.114118.3] # 16, 1 ImplementationClassUID
(0002,0013) SH [eFilm] # 6, 1 ImplementationVersionName
(0002,0016) AE [ ] # 18, 1 SourceApplicationEntityTitle

# Dicom-Data-Set
# Used TransferSyntax: LittleEndianImplicit
(0008,0005) CS [ISO_IR 100] # 10, 1 SpecificCharacterSet
(0008,0008) CS [ORIGINAL\PRIMARY\M_FFE\M\FFE] # 28, 5 ImageType
(0008,0012) DA [20091106] # 8, 1 InstanceCreationDate
(0008,0013) TM [134838] # 6, 1 InstanceCreationTime
(0008,0014) UI [1.3.46.670589.11.17445.5] # 24, 1 InstanceCreatorUID
(0008,0016) UI =MRImageStorage # 26, 1 SOPClassUID
(0008,0018) UI [1.3.46.670589.11.17445.5.0.3368.2009110613452637003] # 52, 1 SOPInstanceUID
(0008,0020) DA [20091106] # 8, 1 StudyDate
(0008,0021) DA [20091106] # 8, 1 SeriesDate
(0008,0022) DA [20091106] # 8, 1 AcquisitionDate
(0008,0023) DA [20091106] # 8, 1 ContentDate
(0008,0030) TM [104438.000000] # 14, 1 StudyTime
(0008,0031) TM [113720.40] # 10, 1 SeriesTime
(0008,0032) TM [113720.40] # 10, 1 AcquisitionTime
(0008,0033) TM [113720.40] # 10, 1 ContentTime
(0008,0050) SH [RA200913006701] # 14, 1 AccessionNumber
(0008,0060) CS [MR] # 2, 1 Modality
(0008,0070) LO [Philips Medical Systems] # 24, 1 Manufacturer
(0008,0100) SH (no value available) # 0, 0 CodeValue
(0008,0102) SH [DCM] # 4, 1 CodingSchemeDesignator
(0008,0104) LO (no value available) # 0, 0 CodeMeaning
(0008,1010) SH [ND1MR] # 6, 1 StationName
(0008,1030) LO [IRM SINUS / MASSIF FACIAL C- C+ -NE] # 36, 1 StudyDescription
(0008,1032) SQ (Sequence with undefined length #=1) # u/l, 1 ProcedureCodeSequence
(fffe,e000) na (Item with undefined length #=5) # u/l, 1 Item
(0008,0000) UL 84 # 4, 1 IdentifyingGroupLength
(0008,0100) SH [8570BSI] # 8, 1 CodeValue
(0008,0102) SH [BROKER] # 6, 1 CodingSchemeDesignator
(0008,0104) LO [IRM SINUS / MASSIF FACIAL C- C+ -NE] # 36, 1 CodeMeaning
(0008,010b) CS [N] # 2, 1 CodeSetExtensionFlag
(fffe,e00d) na (ItemDelimitationItem) # 0, 0 ItemDelimitationItem
(fffe,e0dd) na (SequenceDelimitationItem) # 0, 0 SequenceDelimitationItem
(0008,103e) LO [sWIP THRIVE SUB 10 MIN] # 22, 1 SeriesDescription
(0008,1070) PN (no value available) # 0, 0 OperatorsName
(0008,1080) LO (no value available) # 0, 0 AdmittingDiagnosesDescription
(0008,1090) LO [Achieva] # 8, 1 ManufacturersModelName
(0008,1110) SQ (Sequence with undefined length #=1) # u/l, 1 ReferencedStudySequence
(fffe,e000) na (Item with undefined length #=3) # u/l, 1 Item
(0008,0000) UL 88 # 4, 1 IdentifyingGroupLength
(0008,1150) UI =DetachedStudyManagementSOPClass # 24, 1 ReferencedSOPClassUID
(0008,1155) UI [1.2.124.113532.10.52.2.19.20091106.2558.4312651] # 48, 1 ReferencedSOPInstanceUID
(fffe,e00d) na (ItemDelimitationItem) # 0, 0 ItemDelimitationItem
(fffe,e0dd) na (SequenceDelimitationItem) # 0, 0 SequenceDelimitationItem
(0008,1111) SQ (Sequence with undefined length #=1) # u/l, 1 ReferencedPerformedProcedureStepSequence
(fffe,e000) na (Item with undefined length #=11) # u/l, 1 Item
(0008,0000) UL 154 # 4, 1 IdentifyingGroupLength
(0008,0012) DA [20091106] # 8, 1 InstanceCreationDate
(0008,0013) TM [134801] # 6, 1 InstanceCreationTime
(0008,0014) UI [1.3.46.670589.11.17445.5] # 24, 1 InstanceCreatorUID
(0008,1150) UI =ModalityPerformedProcedureStepSOPClass # 24, 1 ReferencedSOPClassUID
(0008,1155) UI [1.3.46.670589.11.17445.5.0.5356.2009110610443829135] # 52, 1 ReferencedSOPInstanceUID
(0020,0000) UL 10 # 4, 1 ImageGroupLength
(0020,0013) IS [0] # 2, 1 InstanceNumber
(2005,0000) UL 44 # 4, 1 PrivateGroupLength
(2005,0014) LO [Philips MR Imaging DD 005] # 26, 1 PrivateCreator
(2005,1406) ?? 00\00 # 2, 1 Unknown Tag & Data
(fffe,e00d) na (ItemDelimitationItem) # 0, 0 ItemDelimitationItem
(fffe,e0dd) na (SequenceDelimitationItem) # 0, 0 SequenceDelimitationItem
(0008,1120) SQ (Sequence with undefined length #=1) # u/l, 1 ReferencedPatientSequence
(fffe,e000) na (Item with undefined length #=3) # u/l, 1 Item
(0008,0000) UL 92 # 4, 1 IdentifyingGroupLength
(0008,1150) UI =DetachedPatientManagementSOPClass # 24, 1 ReferencedSOPClassUID
(0008,1155) UI [1.2.124.113532.4.64513.48214.20080212.73445.42381053] # 52, 1 ReferencedSOPInstanceUID
(fffe,e00d) na (ItemDelimitationItem) # 0, 0 ItemDelimitationItem
(fffe,e0dd) na (SequenceDelimitationItem) # 0, 0 SequenceDelimitationItem
(0008,1140) SQ (Sequence with undefined length #=3) # u/l, 1 ReferencedImageSequence
(fffe,e000) na (Item with undefined length #=3) # u/l, 1 Item
(0008,0000) UL 94 # 4, 1 IdentifyingGroupLength
(0008,1150) UI =MRImageStorage # 26, 1 ReferencedSOPClassUID
(0008,1155) UI [1.3.46.670589.11.17445.5.0.6904.2009110610513068684] # 52, 1 ReferencedSOPInstanceUID
(fffe,e00d) na (ItemDelimitationItem) # 0, 0 ItemDelimitationItem
(fffe,e000) na (Item with undefined length #=3) # u/l, 1 Item
(0008,0000) UL 94 # 4, 1 IdentifyingGroupLength
(0008,1150) UI =MRImageStorage # 26, 1 ReferencedSOPClassUID
(0008,1155) UI [1.3.46.670589.11.17445.5.0.8024.2009110610465923028] # 52, 1 ReferencedSOPInstanceUID
(fffe,e00d) na (ItemDelimitationItem) # 0, 0 ItemDelimitationItem
(fffe,e000) na (Item with undefined length #=3) # u/l, 1 Item
(0008,0000) UL 94 # 4, 1 IdentifyingGroupLength
(0008,1150) UI =MRImageStorage # 26, 1 ReferencedSOPClassUID
(0008,1155) UI [1.3.46.670589.11.17445.5.0.6904.2009110610584771735] # 52, 1 ReferencedSOPInstanceUID
(fffe,e00d) na (ItemDelimitationItem) # 0, 0 ItemDelimitationItem
(fffe,e0dd) na (SequenceDelimitationItem) # 0, 0 SequenceDelimitationItem
(0010,0000) UL 126 # 4, 1 PatientGroupLength
(0010,1030) DS [55] # 2, 1 PatientsWeight
(0010,21c0) US 4 # 2, 1 PregnancyStatus
(0018,0000) UL 566 # 4, 1 AcquisitionGroupLength
(0018,0015) CS [HEAD] # 4, 1 BodyPartExamined
(0018,0020) CS [GR] # 2, 1 ScanningSequence
(0018,0021) CS [SP] # 2, 1 SequenceVariant
(0018,0022) CS [FS] # 2, 1 ScanOptions
(0018,0023) CS [3D] # 2, 1 MRAcquisitionType
(0018,0024) SH (no value available) # 0, 0 SequenceName
(0018,0050) DS [1.5] # 4, 1 SliceThickness
(0018,0080) DS [5.92609977722168] # 16, 1 RepetitionTime
(0018,0081) DS [2.732] # 6, 1 EchoTime
(0018,0083) DS [1] # 2, 1 NumberOfAverages
(0018,0084) DS [127.802354999999] # 16, 1 ImagingFrequency
(0018,0085) SH [1H] # 2, 1 ImagedNucleus
(0018,0086) IS [1] # 2, 1 EchoNumbers
(0018,0087) DS [3] # 2, 1 MagneticFieldStrength
(0018,0088) DS [1.5] # 4, 1 SpacingBetweenSlices
(0018,0089) IS [99] # 2, 1 NumberOfPhaseEncodingSteps
(0018,0091) IS [30] # 2, 1 EchoTrainLength
(0018,0093) DS [77.7544174194336] # 16, 1 PercentSampling
(0018,0094) DS [125] # 4, 1 PercentPhaseFieldOfView
(0018,0095) DS [287] # 4, 1 PixelBandwidth
(0018,1000) LO [17445] # 6, 1 DeviceSerialNumber
(0018,1020) LO [2.6.3] # 6, 1 SoftwareVersions
(0018,1030) LO [sWIP THRIVE SUB 10 MIN] # 22, 1 ProtocolName
(0018,1081) IS [0] # 2, 1 LowRRValue
(0018,1082) IS [0] # 2, 1 HighRRValue
(0018,1083) IS [0] # 2, 1 IntervalsAcquired
(0018,1084) IS [0] # 2, 1 IntervalsRejected
(0018,1088) IS [0] # 2, 1 HeartRate
(0018,1100) DS [150] # 4, 1 ReconstructionDiameter
(0018,1250) SH [SENSE-NV-16] # 12, 1 ReceiveCoilName
(0018,1251) SH [B] # 2, 1 TransmitCoilName
(0018,1310) US 0\100\99\0 # 8, 4 AcquisitionMatrix
(0018,1312) CS [ROW] # 4, 1 InPlanePhaseEncodingDirection
(0018,1314) DS [7] # 2, 1 FlipAngle
(0018,1316) DS [0.94299358129501] # 16, 1 SAR
(0018,1318) DS [51.1089059627999] # 16, 1 dBdt
(0018,5100) CS [HFS] # 4, 1 PatientPosition
(0018,9073) FD 59.882 # 8, 1 AcquisitionDuration
(0018,9087) FD 0 # 8, 1 DiffusionBValue
(0018,9089) FD 0\0\0 # 24, 3 DiffusionGradientOrientation
(0020,0000) UL 450 # 4, 1 ImageGroupLength
(0020,000d) UI [1.2.124.113532.4.64513.48214.20091106.2616.47986432] # 52, 1 StudyInstanceUID
(0020,000e) UI [1.3.46.670589.11.17445.5.0.3368.2009110613452396000] # 52, 1 SeriesInstanceUID
(0020,0010) SH [RA200913006701] # 14, 1 StudyID
(0020,0011) IS [1702] # 4, 1 SeriesNumber
(0020,0012) IS [17] # 2, 1 AcquisitionNumber
(0020,0013) IS [1] # 2, 1 InstanceNumber
(0020,0032) DS [-83.640634685754\-143.61624196916\-62.057746164500] # 50, 3 ImagePositionPatient
(0020,0037) DS [0.99714380502700\0.03825115412473\-0.0651234090328\-0.043147653341... # 102, 6 ImageOrientationPatient
(0020,0052) UI [1.3.46.670589.11.17445.5.0.4480.2009110610435484005] # 52, 1 FrameOfReferenceUID
(0020,0060) CS [L] # 2, 1 Laterality
(0020,0100) IS [1] # 2, 1 TemporalPositionIdentifier
(0020,0105) IS [1] # 2, 1 NumberOfTemporalPositions
(0020,1040) LO (no value available) # 0, 0 PositionReferenceIndicator
(0020,1041) DS [0] # 2, 1 SliceLocation
(0028,0000) UL 146 # 4, 1 ImagePresentationGroupLength
(0028,0002) US 1 # 2, 1 SamplesPerPixel
(0028,0004) CS [MONOCHROME2] # 12, 1 PhotometricInterpretation
(0028,0010) US 400 # 2, 1 Rows
(0028,0011) US 400 # 2, 1 Columns
(0028,0030) DS [0.375\0.375] # 12, 2 PixelSpacing
(0028,0100) US 16 # 2, 1 BitsAllocated
(0028,0101) US 12 # 2, 1 BitsStored
(0028,0102) US 11 # 2, 1 HighBit
(0028,0103) US 0 # 2, 1 PixelRepresentation
(0028,1050) DS [2047.5] # 6, 1 WindowCenter
(0028,1051) DS [4095] # 4, 1 WindowWidth
(0028,2110) CS [00] # 2, 1 LossyImageCompression
(0031,0000) UL 118 # 4, 1 PrivateGroupLength
(0031,0010) LO [MITRA LINKED ATTRIBUTES 1.0] # 28, 1 PrivateCreator
(0031,0011) LO [AGFA PACS Archive Mirroring 1.0] # 32, 1 PrivateCreator
(0031,1020) ?? 50\41\4e\41\36\30\35\31\30\34\31\38 # 12, 1 Unknown Tag & Data
(0031,1100) ?? 52\20 # 2, 1 Unknown Tag & Data
(0031,1101) ?? 09\b1\1d\4b # 4, 1 Unknown Tag & Data
(0032,0000) UL 100 # 4, 1 StudyGroupLength
(0032,000a) CS [READ] # 4, 1 StudyStatusID
(0032,1060) LO [IRM SINUS / MASSIF FACIAL C- C+ -NE] # 36, 1 RequestedProcedureDescription
(0033,0000) UL 176 # 4, 1 PrivateGroupLength
(0033,0010) LO [MITRA OBJECT UTF8 ATTRIBUTES 1.0] # 32, 1 PrivateCreator
(0033,1004) ?? 49\52\4d\20\53\49\4e\55\53\20\2f\20\4d\41\53\53\49\46\20\46\41\43... # 36, 1 Unknown Tag & Data
(0033,1013) ?? 50\41\4e\49\43\43\49\41\5e\41\4e\47\45\4c\41\20\4d\41\52\49\54\5a... # 24, 1 Unknown Tag & Data
(0033,1016) ?? 39\35\33\34\36\5e\43\6f\74\65\2c\4d\61\6e\6f\6e\5e\5e\5e\20 # 20, 1 Unknown Tag & Data
(0033,1019) ?? 31\39\31\32\36\32\5e\53\4f\55\4c\45\5a\2c\47\49\4c\4c\45\53\5e\5e... # 24, 1 Unknown Tag & Data
(0040,0000) UL 200 # 4, 1 ModalityWorklistGroupLength
(0040,0241) AE [ND1MR] # 6, 1 PerformedStationAETitle
(0040,0244) DA [20091106] # 8, 1 PerformedProcedureStepStartDate
(0040,0245) TM [104438] # 6, 1 PerformedProcedureStepStartTime
(0040,0250) DA [20091106] # 8, 1 PerformedProcedureStepEndDate
(0040,0251) TM [104438] # 6, 1 PerformedProcedureStepEndTime
(0040,0253) SH [310646678] # 10, 1 PerformedProcedureStepID
(0040,0254) LO [IRM SINUS / MASSIF FACIAL C- C+ -NE] # 36, 1 PerformedProcedureStepDescription
(0040,0260) SQ (Sequence with undefined length #=1) # u/l, 1 PerformedProtocolCodeSequence
(fffe,e000) na (Item with undefined length #=5) # u/l, 1 Item
(0008,0000) UL 84 # 4, 1 IdentifyingGroupLength
(0008,0100) SH [8570BSI] # 8, 1 CodeValue
(0008,0102) SH [BROKER] # 6, 1 CodingSchemeDesignator
(0008,0104) LO [IRM SINUS / MASSIF FACIAL C- C+ -NE] # 36, 1 CodeMeaning
(0008,010b) CS [N] # 2, 1 CodeSetExtensionFlag
(fffe,e00d) na (ItemDelimitationItem) # 0, 0 ItemDelimitationItem
(fffe,e0dd) na (SequenceDelimitationItem) # 0, 0 SequenceDelimitationItem
(0040,0275) SQ (Sequence with undefined length #=1) # u/l, 1 RequestAttributesSequence
(fffe,e000) na (Item with undefined length #=3) # u/l, 1 Item
(0040,0000) UL 44 # 4, 1 ModalityWorklistGroupLength
(0040,0009) SH [RA200913006701] # 14, 1 ScheduledProcedureStepID
(0040,1001) SH [RA200913006701] # 14, 1 RequestedProcedureID
(fffe,e00d) na (ItemDelimitationItem) # 0, 0 ItemDelimitationItem
(fffe,e0dd) na (SequenceDelimitationItem) # 0, 0 SequenceDelimitationItem
(0040,0321) SQ (Sequence with explicit length #=0) # 0, 1 FilmConsumptionSequence
(fffe,e0dd) na (SequenceDelimitationItem for re-encod.) # 0, 0 SequenceDelimitationItem
(0040,1001) SH [RA200913006701] # 14, 1 RequestedProcedureID
(0040,2400) LT [] # 2, 0 ImagingServiceRequestComments
(0040,9096) SQ (Sequence with undefined length #=1) # u/l, 1 RealWorldValueMappingSequence
(fffe,e000) na (Item with undefined length #=3) # u/l, 1 Item
(0040,0000) UL 32 # 4, 1 ModalityWorklistGroupLength
(0040,9224) FD 0 # 8, 1 RealWorldValueIntercept
(0040,9225) FD 1 # 8, 1 RealWorldValueSlope
(fffe,e00d) na (ItemDelimitationItem) # 0, 0 ItemDelimitationItem
(fffe,e0dd) na (SequenceDelimitationItem) # 0, 0 SequenceDelimitationItem
(2001,0000) UL 668 # 4, 1 PrivateGroupLength
(2001,0010) LO [Philips Imaging DD 001] # 22, 1 PrivateCreator
(2001,0090) LO [Philips Imaging DD 129] # 22, 1 PrivateCreator
(2001,1001) ?? 00\00\00\00 # 4, 1 Unknown Tag & Data
(2001,1002) ?? 30\20 # 2, 1 Unknown Tag & Data
(2001,1003) ?? 00\00\00\00 # 4, 1 Unknown Tag & Data
(2001,1006) ?? 4e\20 # 2, 1 Unknown Tag & Data
(2001,1007) ?? 55\20 # 2, 1 Unknown Tag & Data
(2001,1008) ?? 31\20 # 2, 1 Unknown Tag & Data
(2001,1009) ?? 00\00\00\00 # 4, 1 Unknown Tag & Data
(2001,100a) ?? 31\20 # 2, 1 Unknown Tag & Data
(2001,100b) ?? 54\52\41\4e\53\56\45\52\53\41\4c\20 # 12, 1 Unknown Tag & Data
(2001,100c) ?? 4e\20 # 2, 1 Unknown Tag & Data
(2001,100e) ?? 4e\20 # 2, 1 Unknown Tag & Data
(2001,100f) ?? 00\00 # 2, 1 Unknown Tag & Data
(2001,1010) ?? 4e\4f # 2, 1 Unknown Tag & Data
(2001,1011) ?? 00\00\00\00 # 4, 1 Unknown Tag & Data
(2001,1012) ?? 4e\20 # 2, 1 Unknown Tag & Data
(2001,1013) ?? 01\00\00\00 # 4, 1 Unknown Tag & Data
(2001,1014) ?? 01\00\00\00 # 4, 1 Unknown Tag & Data
(2001,1015) ?? 01\00 # 2, 1 Unknown Tag & Data
(2001,1016) ?? 00\00 # 2, 1 Unknown Tag & Data
(2001,1017) ?? 01\00\00\00 # 4, 1 Unknown Tag & Data
(2001,1018) ?? 32\00\00\00 # 4, 1 Unknown Tag & Data
(2001,1019) ?? 4e\20 # 2, 1 Unknown Tag & Data
(2001,101a) ?? 00\00\00\00\00\00\00\00\00\00\00\00 # 12, 1 Unknown Tag & Data
(2001,101b) ?? 00\00\00\00 # 4, 1 Unknown Tag & Data
(2001,101c) ?? 4e\4f # 2, 1 Unknown Tag & Data
(2001,101d) ?? 32\20 # 2, 1 Unknown Tag & Data
(2001,101f) ?? 4e\4f # 2, 1 Unknown Tag & Data
(2001,1020) ?? 54\31\54\46\45\20 # 6, 1 Unknown Tag & Data
(2001,1021) ?? 59\20 # 2, 1 Unknown Tag & Data
(2001,1022) ?? 36\c0\c1\3f # 4, 1 Unknown Tag & Data
(2001,1023) ?? 37\20 # 2, 1 Unknown Tag & Data
(2001,1024) ?? 4e\20 # 2, 1 Unknown Tag & Data
(2001,1025) ?? 32\2e\37\20 # 4, 1 Unknown Tag & Data
(2001,105f) SQ (Sequence with undefined length #=1) # u/l, 1 Unknown Tag & Data
(fffe,e000) na (Item with undefined length #=32) # u/l, 1 Item
(2001,0000) UL 88 # 4, 1 PrivateGroupLength
(2001,0010) LO [Philips Imaging DD 001] # 22, 1 PrivateCreator
(2001,102d) ?? 32\00 # 2, 1 Unknown Tag & Data
(2001,1032) ?? 00\00\00\00 # 4, 1 Unknown Tag & Data
(2001,1033) ?? 41\50 # 2, 1 Unknown Tag & Data
(2001,1035) ?? 01\00 # 2, 1 Unknown Tag & Data
(2001,1036) ?? 50\41\52\41\4c\4c\45\4c # 8, 1 Unknown Tag & Data
(2005,0000) UL 346 # 4, 1 PrivateGroupLength
(2005,0010) LO [Philips MR Imaging DD 001] # 26, 1 PrivateCreator
(2005,0013) LO [Philips MR Imaging DD 004] # 26, 1 PrivateCreator
(2005,0014) LO [Philips MR Imaging DD 005] # 26, 1 PrivateCreator
(2005,1071) ?? 67\73\63\40 # 4, 1 Unknown Tag & Data
(2005,1072) ?? de\92\1e\40 # 4, 1 Unknown Tag & Data
(2005,1073) ?? 43\a5\8f\c0 # 4, 1 Unknown Tag & Data
(2005,1074) ?? 00\00\f0\42 # 4, 1 Unknown Tag & Data
(2005,1075) ?? 00\00\96\42 # 4, 1 Unknown Tag & Data
(2005,1076) ?? 00\00\16\43 # 4, 1 Unknown Tag & Data
(2005,1078) ?? c8\a4\7c\c2 # 4, 1 Unknown Tag & Data
(2005,1079) ?? 40\28\10\c2 # 4, 1 Unknown Tag & Data
(2005,107a) ?? 76\01\1d\c1 # 4, 1 Unknown Tag & Data
(2005,107b) ?? 52\4c # 2, 1 Unknown Tag & Data
(2005,107e) ?? 00\00\c0\3f # 4, 1 Unknown Tag & Data
(2005,1081) ?? 46\48 # 2, 1 Unknown Tag & Data
(2005,10a3) ?? 36\37 # 2, 1 Unknown Tag & Data
(2005,10a4) ?? 30\20 # 2, 1 Unknown Tag & Data
(2005,10a5) ?? 30\20 # 2, 1 Unknown Tag & Data
(2005,10a6) ?? 36\33 # 2, 1 Unknown Tag & Data
(2005,10a7) ?? 44\45\46\41\55\4c\54\20 # 8, 1 Unknown Tag & Data
(2005,1390) ?? 52\5f\41\42 # 4, 1 Unknown Tag & Data
(2005,143c) ?? 9e\c9\ff\7e # 4, 1 Unknown Tag & Data
(2005,143d) ?? 9e\c9\ff\7e # 4, 1 Unknown Tag & Data
(2005,143e) ?? 9e\c9\ff\7e # 4, 1 Unknown Tag & Data
(fffe,e00d) na (ItemDelimitationItem) # 0, 0 ItemDelimitationItem
(fffe,e0dd) na (SequenceDelimitationItem) # 0, 0 SequenceDelimitationItem
(2001,1060) ?? 01\00\00\00 # 4, 1 Unknown Tag & Data
(2001,1061) ?? 4e\20 # 2, 1 Unknown Tag & Data
(2001,1062) ?? 4e\20 # 2, 1 Unknown Tag & Data
(2001,1063) ?? 45\4c\53\45\57\48\45\52\45\20 # 10, 1 Unknown Tag & Data
(2001,107b) ?? 31\37 # 2, 1 Unknown Tag & Data
(2001,1081) ?? 31\20 # 2, 1 Unknown Tag & Data
(2001,1082) ?? 33\30 # 2, 1 Unknown Tag & Data
(2001,1083) ?? 31\32\37\2e\38\30\32\33\35\34\39\39\39\39\39\39 # 16, 1 Unknown Tag & Data
(2001,1084) ?? 30\20 # 2, 1 Unknown Tag & Data
(2001,1085) ?? 33\20 # 2, 1 Unknown Tag & Data
(2001,1086) ?? 30\20 # 2, 1 Unknown Tag & Data
(2001,1087) ?? 31\48 # 2, 1 Unknown Tag & Data
(2001,1088) ?? 31\20 # 2, 1 Unknown Tag & Data
(2001,1089) ?? 30\20 # 2, 1 Unknown Tag & Data
(2001,108a) ?? 30\20 # 2, 1 Unknown Tag & Data
(2001,108b) ?? 42\20 # 2, 1 Unknown Tag & Data
(2001,10f1) ?? 00\00\00\00\00\00\00\00\00\00\00\00\00\00\00\00\00\00\00\00\00\00... # 24, 1 Unknown Tag & Data
(2001,9000) SQ (Sequence with undefined length #=1) # u/l, 1 Unknown Tag & Data
(fffe,e000) na (Item with undefined length #=43) # u/l, 1 Item
(0008,0000) UL 446 # 4, 1 IdentifyingGroupLength
(0008,0005) CS [ISO_IR 100] # 10, 1 SpecificCharacterSet
(0008,0012) DA [20091106] # 8, 1 InstanceCreationDate
(0008,0013) TM [134838] # 6, 1 InstanceCreationTime
(0008,0014) UI [1.3.46.670589.11.17445.5] # 24, 1 InstanceCreatorUID
(0008,0016) UI =GrayscaleSoftcopyPresentationStateStorage # 28, 1 SOPClassUID
(0008,0018) UI [1.3.46.670589.11.17445.5.0.3368.2009110613452398001] # 52, 1 SOPInstanceUID
(0008,1115) SQ (Sequence with undefined length #=1) # u/l, 1 ReferencedSeriesSequence
(fffe,e000) na (Item with undefined length #=4) # u/l, 1 Item
(0008,0000) UL 146 # 4, 1 IdentifyingGroupLength
(0008,1140) SQ (Sequence with undefined length #=1) # u/l, 1 ReferencedImageSequence
(fffe,e000) na (Item with undefined length #=3) # u/l, 1 Item
(0008,0000) UL 94 # 4, 1 IdentifyingGroupLength
(0008,1150) UI =MRImageStorage # 26, 1 ReferencedSOPClassUID
(0008,1155) UI [1.3.46.670589.11.17445.5.0.3368.2009110613452637003] # 52, 1 ReferencedSOPInstanceUID
(fffe,e00d) na (ItemDelimitationItem) # 0, 0 ItemDelimitationItem
(fffe,e0dd) na (SequenceDelimitationItem) # 0, 0 SequenceDelimitationItem
(0020,0000) UL 60 # 4, 1 ImageGroupLength
(0020,000e) UI [1.3.46.670589.11.17445.5.0.3368.2009110613452396000] # 52, 1 SeriesInstanceUID
(fffe,e00d) na (ItemDelimitationItem) # 0, 0 ItemDelimitationItem
(fffe,e0dd) na (SequenceDelimitationItem) # 0, 0 SequenceDelimitationItem
(0018,0000) UL 10 # 4, 1 AcquisitionGroupLength
(0018,1622) US 0 # 2, 1 ShutterPresentationValue
(0020,0000) UL 10 # 4, 1 ImageGroupLength
(0020,0013) IS [0] # 2, 1 InstanceNumber
(0028,0000) UL 138 # 4, 1 ImagePresentationGroupLength
(0028,3110) SQ (Sequence with undefined length #=1) # u/l, 1 SoftcopyVOILUTSequence
(fffe,e000) na (Item with undefined length #=6) # u/l, 1 Item
(0028,0000) UL 26 # 4, 1 ImagePresentationGroupLength
(0028,1050) DS [2047.5] # 6, 1 WindowCenter
(0028,1051) DS [4095] # 4, 1 WindowWidth
(2001,0000) UL 48 # 4, 1 PrivateGroupLength
(2001,0010) LO [Philips Imaging DD 001] # 22, 1 PrivateCreator
(2001,1077) ?? 4c\49\4e\45\41\52\56\4f\49\20 # 10, 1 Unknown Tag & Data
(fffe,e00d) na (ItemDelimitationItem) # 0, 0 ItemDelimitationItem
(fffe,e0dd) na (SequenceDelimitationItem) # 0, 0 SequenceDelimitationItem
(0070,0000) UL 324 # 4, 1 PresentationStateGroupLength
(0070,005a) SQ (Sequence with undefined length #=1) # u/l, 1 DisplayedAreaSelectionSequence
(fffe,e000) na (Item with undefined length #=8) # u/l, 1 Item
(0070,0000) UL 72 # 4, 1 PresentationStateGroupLength
(0070,0052) SL 1\1 # 8, 2 DisplayedAreaTopLeftHandCorner
(0070,0053) SL 400\400 # 8, 2 DisplayedAreaBottomRightHandCorner
(0070,0100) CS [SCALE TO FIT] # 12, 1 PresentationSizeMode
(0070,0101) DS [0.375\0.375] # 12, 2 PresentationPixelSpacing
(2001,0000) UL 46 # 4, 1 PrivateGroupLength
(2001,0010) LO [Philips Imaging DD 001] # 22, 1 PrivateCreator
(2001,103f) ?? 42\49\5f\43\55\42\49\43 # 8, 1 Unknown Tag & Data
(fffe,e00d) na (ItemDelimitationItem) # 0, 0 ItemDelimitationItem
(fffe,e0dd) na (SequenceDelimitationItem) # 0, 0 SequenceDelimitationItem
(0070,0060) SQ (Sequence with undefined length #=1) # u/l, 1 GraphicLayerSequence
(fffe,e000) na (Item with undefined length #=3) # u/l, 1 Item
(0070,0000) UL 20 # 4, 1 PresentationStateGroupLength
(0070,0002) CS [1] # 2, 1 GraphicLayer
(0070,0062) IS [1] # 2, 1 GraphicLayerOrder
(fffe,e00d) na (ItemDelimitationItem) # 0, 0 ItemDelimitationItem
(fffe,e0dd) na (SequenceDelimitationItem) # 0, 0 SequenceDelimitationItem
(0070,0080) CS [AS LAST SEEN] # 12, 1 ContentLabel
(0070,0081) LO (no value available) # 0, 0 ContentDescription
(0070,0082) DA [20091106] # 8, 1 PresentationCreationDate
(0070,0083) TM [134523.984] # 10, 1 PresentationCreationTime
(0070,0084) PN (no value available) # 0, 0 ContentCreatorsName
(2001,0000) UL 274 # 4, 1 PrivateGroupLength
(2001,0010) LO [Philips Imaging DD 001] # 22, 1 PrivateCreator
(2001,0090) LO [Philips Imaging DD 129] # 22, 1 PrivateCreator
(2001,1026) ?? 4e\20 # 2, 1 Unknown Tag & Data
(2001,1067) ?? (no value available) # 0, 1 Unknown Tag & Data
(2001,1068) SQ (Sequence with undefined length #=1) # u/l, 1 Unknown Tag & Data
(fffe,e000) na (Item with undefined length #=7) # u/l, 1 Item
(0028,0000) UL 38 # 4, 1 ImagePresentationGroupLength
(0028,1052) DS [0] # 2, 1 RescaleIntercept
(0028,1053) DS [1] # 2, 1 RescaleSlope
(0028,1054) LO [normalized] # 10, 1 RescaleType
(2001,0000) UL 52 # 4, 1 PrivateGroupLength
(2001,0010) LO [Philips Imaging DD 001] # 22, 1 PrivateCreator
(2001,1077) ?? 4c\49\4e\45\41\52\4d\4f\44\41\4c\49\54\59 # 14, 1 Unknown Tag & Data
(fffe,e00d) na (ItemDelimitationItem) # 0, 0 ItemDelimitationItem
(fffe,e0dd) na (SequenceDelimitationItem) # 0, 0 SequenceDelimitationItem
(2001,1077) ?? 4c\5f\50\52\45\53\45\4e\54\41\54\49\4f\4e # 14, 1 Unknown Tag & Data
(2001,109f) ?? 00\00\00\00 # 4, 1 Unknown Tag & Data
(2001,9000) ?? (no value available) # 0, 1 Unknown Tag & Data
(2005,0000) UL 158 # 4, 1 PrivateGroupLength
(2005,0012) LO [Philips MR Imaging DD 003] # 26, 1 PrivateCreator
(2005,0013) LO [Philips MR Imaging DD 004] # 26, 1 PrivateCreator
(2005,0014) LO [Philips MR Imaging DD 005] # 26, 1 PrivateCreator
(2005,1234) ?? 00\00\00\00 # 4, 1 Unknown Tag & Data
(2005,1257) ?? 00\00 # 2, 1 Unknown Tag & Data
(2005,1281) ?? (no value available) # 0, 1 Unknown Tag & Data
(2005,1282) ?? (no value available) # 0, 1 Unknown Tag & Data
(2005,1389) ?? (no value available) # 0, 1 Unknown Tag & Data
(2005,1406) ?? 00\00 # 2, 1 Unknown Tag & Data
(2050,0000) UL 16 # 4, 1 PresentationLUTGroupLength
(2050,0020) CS [IDENTITY] # 8, 1 PresentationLUTShape
(fffe,e00d) na (ItemDelimitationItem) # 0, 0 ItemDelimitationItem
(fffe,e0dd) na (SequenceDelimitationItem) # 0, 0 SequenceDelimitationItem
(2005,0000) UL 2936 # 4, 1 PrivateGroupLength
(2005,0010) LO [Philips MR Imaging DD 001] # 26, 1 PrivateCreator
(2005,0011) LO [Philips MR Imaging DD 002] # 26, 1 PrivateCreator
(2005,0012) LO [Philips MR Imaging DD 003] # 26, 1 PrivateCreator
(2005,0013) LO [Philips MR Imaging DD 004] # 26, 1 PrivateCreator
(2005,0014) LO [Philips MR Imaging DD 005] # 26, 1 PrivateCreator
(2005,1000) ?? 67\73\63\40 # 4, 1 Unknown Tag & Data
(2005,1001) ?? de\92\1e\40 # 4, 1 Unknown Tag & Data
(2005,1002) ?? 43\a5\8f\c0 # 4, 1 Unknown Tag & Data
(2005,1004) ?? 4e\4f\4e\45 # 4, 1 Unknown Tag & Data
(2005,1008) ?? 33\10\84\c2 # 4, 1 Unknown Tag & Data
(2005,1009) ?? 56\36\91\c2 # 4, 1 Unknown Tag & Data
(2005,100a) ?? 6c\74\41\c1 # 4, 1 Unknown Tag & Data
(2005,100b) ?? 2f\a5\4f\48 # 4, 1 Unknown Tag & Data
(2005,100c) ?? 00\00\00\00 # 4, 1 Unknown Tag & Data
(2005,100d) ?? 52\a8\ae\44 # 4, 1 Unknown Tag & Data
(2005,100e) ?? e2\a4\09\3c # 4, 1 Unknown Tag & Data
(2005,100f) ?? 32\30\34\37\2e\35 # 6, 1 Unknown Tag & Data
(2005,1010) ?? 34\30\39\35 # 4, 1 Unknown Tag & Data
(2005,1011) ?? 4d\20 # 2, 1 Unknown Tag & Data
(2005,1012) ?? 4e\20 # 2, 1 Unknown Tag & Data
(2005,1013) ?? 50\52\4f\54\4f\20 # 6, 1 Unknown Tag & Data
(2005,1014) ?? 4e\20 # 2, 1 Unknown Tag & Data
(2005,1015) ?? 59\20 # 2, 1 Unknown Tag & Data
(2005,1016) ?? 4e\20 # 2, 1 Unknown Tag & Data
(2005,1017) ?? 4e\20 # 2, 1 Unknown Tag & Data
(2005,1019) ?? 4e\20 # 2, 1 Unknown Tag & Data
(2005,101a) ?? 00\00 # 2, 1 Unknown Tag & Data
(2005,101b) ?? 4e\20 # 2, 1 Unknown Tag & Data
(2005,101c) ?? 4e\20 # 2, 1 Unknown Tag & Data
(2005,101d) ?? 64\00 # 2, 1 Unknown Tag & Data
(2005,101e) ?? 63\6f\6d\70\6f\73\65\20 # 8, 1 Unknown Tag & Data
(2005,101f) ?? 63\6f\6d\70\6f\73\65\20 # 8, 1 Unknown Tag & Data
(2005,1020) ?? 00\00\00\00 # 4, 1 Unknown Tag & Data
(2005,1021) ?? 01\00 # 2, 1 Unknown Tag & Data
(2005,1022) ?? 31\20 # 2, 1 Unknown Tag & Data
(2005,1023) ?? 00\00 # 2, 1 Unknown Tag & Data
(2005,1025) ?? 00\00 # 2, 1 Unknown Tag & Data
(2005,1026) ?? 4e\20 # 2, 1 Unknown Tag & Data
(2005,1027) ?? 4d\41\58\49\4d\55\4d\20 # 8, 1 Unknown Tag & Data
(2005,1028) ?? 4e\20 # 2, 1 Unknown Tag & Data
(2005,1029) ?? 4e\20 # 2, 1 Unknown Tag & Data
(2005,102a) ?? 33\31\39\32\38\36\36\34\34\20 # 10, 1 Unknown Tag & Data
(2005,102b) ?? 64\00 # 2, 1 Unknown Tag & Data
(2005,102c) ?? 4e\20 # 2, 1 Unknown Tag & Data
(2005,102d) ?? 30\20 # 2, 1 Unknown Tag & Data
(2005,102e) ?? 4e\20 # 2, 1 Unknown Tag & Data
(2005,102f) ?? 4e\20 # 2, 1 Unknown Tag & Data
(2005,1030) ?? 9c\a2\bd\40 # 4, 1 Unknown Tag & Data
(2005,1031) ?? 4e\20 # 2, 1 Unknown Tag & Data
(2005,1033) ?? 2b\87\6f\42 # 4, 1 Unknown Tag & Data
(2005,1034) ?? 59\20 # 2, 1 Unknown Tag & Data
(2005,1035) ?? 50\49\58\45\4c\20 # 6, 1 Unknown Tag & Data
(2005,1036) ?? 4e\20 # 2, 1 Unknown Tag & Data
(2005,1037) ?? 4e\20 # 2, 1 Unknown Tag & Data
(2005,1038) ?? 59\20 # 2, 1 Unknown Tag & Data
(2005,1039) ?? 4e\20 # 2, 1 Unknown Tag & Data
(2005,103b) ?? 4e\20 # 2, 1 Unknown Tag & Data
(2005,103c) ?? 4e\20 # 2, 1 Unknown Tag & Data
(2005,103d) ?? 66\00 # 2, 1 Unknown Tag & Data
(2005,103e) ?? 00\00\00\00\00\00\00\00\00\00\00\00\00\00\00\00\00\00\00\00\00\00... # 408, 1 Unknown Tag & Data
(2005,105f) ?? 55\4e\4b\4e\4f\57\4e\20 # 8, 1 Unknown Tag & Data
(2005,1060) ?? 2d\31 # 2, 1 Unknown Tag & Data
(2005,1061) ?? 4e\4f # 2, 1 Unknown Tag & Data
(2005,1063) ?? 00\00 # 2, 1 Unknown Tag & Data
(2005,106e) ?? 46\46\45\20 # 4, 1 Unknown Tag & Data
(2005,106f) ?? 33\44 # 2, 1 Unknown Tag & Data
(2005,1080) ?? (no value available) # 0, 1 Unknown Tag & Data
(2005,1083) ?? (no value available) # 0, 1 Unknown Tag & Data
(2005,1084) SQ (Sequence with undefined length #=1) # u/l, 1 Unknown Tag & Data
(fffe,e000) na (Item with undefined length #=15) # u/l, 1 Item
(2005,0000) UL 196 # 4, 1 PrivateGroupLength
(2005,0010) LO [Philips MR Imaging DD 001] # 26, 1 PrivateCreator
(2005,0014) LO [Philips MR Imaging DD 005] # 26, 1 PrivateCreator
(2005,1048) ?? 31\37 # 2, 1 Unknown Tag & Data
(2005,1049) ?? 30\20 # 2, 1 Unknown Tag & Data
(2005,104a) ?? 30\20 # 2, 1 Unknown Tag & Data
(2005,104b) ?? 30\20 # 2, 1 Unknown Tag & Data
(2005,104c) ?? 53\45\52\49\45\53 # 6, 1 Unknown Tag & Data
(2005,104d) ?? (no value available) # 0, 1 Unknown Tag & Data
(2005,1064) ?? 30\20 # 2, 1 Unknown Tag & Data
(2005,1065) ?? 31\20 # 2, 1 Unknown Tag & Data
(2005,1066) ?? (no value available) # 0, 1 Unknown Tag & Data
(2005,1067) ?? 30\20 # 2, 1 Unknown Tag & Data
(2005,1068) ?? 49\4d\41\47\45\41\4c\47\45\42\52\41 # 12, 1 Unknown Tag & Data
(2005,1436) ?? (no value available) # 0, 1 Unknown Tag & Data
(fffe,e00d) na (ItemDelimitationItem) # 0, 0 ItemDelimitationItem
(fffe,e0dd) na (SequenceDelimitationItem) # 0, 0 SequenceDelimitationItem
(2005,1085) ?? (no value available) # 0, 1 Unknown Tag & Data
(2005,1086) ?? 00\00 # 2, 1 Unknown Tag & Data
(2005,109e) ?? (no value available) # 0, 1 Unknown Tag & Data
(2005,10a0) ?? 00\00\00\00 # 4, 1 Unknown Tag & Data
(2005,10a1) ?? 53\59\4e\5f\43\4f\43\41 # 8, 1 Unknown Tag & Data
(2005,10a2) ?? 4e\20 # 2, 1 Unknown Tag & Data
(2005,10a8) ?? 30\20 # 2, 1 Unknown Tag & Data
(2005,10a9) ?? 33\44 # 2, 1 Unknown Tag & Data
(2005,10b0) ?? 00\00\00\00 # 4, 1 Unknown Tag & Data
(2005,10b1) ?? 00\00\00\00 # 4, 1 Unknown Tag & Data
(2005,10b2) ?? 00\00\00\00 # 4, 1 Unknown Tag & Data
(2005,10c0) ?? 47\52 # 2, 1 Unknown Tag & Data
(2005,1199) ?? 01\00\00\00 # 4, 1 Unknown Tag & Data
(2005,1200) ?? 01\00\00\00 # 4, 1 Unknown Tag & Data
(2005,1201) ?? 00\00\00\00 # 4, 1 Unknown Tag & Data
(2005,1213) ?? 01\00\00\00 # 4, 1 Unknown Tag & Data
(2005,1245) ?? 01\00 # 2, 1 Unknown Tag & Data
(2005,1249) ?? 00\00 # 2, 1 Unknown Tag & Data
(2005,1251) ?? 00\00 # 2, 1 Unknown Tag & Data
(2005,1252) ?? 01\00 # 2, 1 Unknown Tag & Data
(2005,1253) ?? 01\00 # 2, 1 Unknown Tag & Data
(2005,1325) ?? 4e\20 # 2, 1 Unknown Tag & Data
(2005,1326) ?? 00\00\00\00 # 4, 1 Unknown Tag & Data
(2005,1327) ?? 52\45\41\4c # 4, 1 Unknown Tag & Data
(2005,1328) ?? 4f\52\49\47\49\4e\41\4c # 8, 1 Unknown Tag & Data
(2005,1329) ?? 00\00\48\42 # 4, 1 Unknown Tag & Data
(2005,1330) ?? 50\4c\55\53\5f\41\5f\50\4c\55\53\5f\42\5c\50\4c\55\53\5f\41\5f\50... # 56, 1 Unknown Tag & Data
(2005,1331) ?? 00\00 # 2, 1 Unknown Tag & Data
(2005,1333) ?? 00\00\00\00\00\00\00\00\00\00\00\00 # 12, 1 Unknown Tag & Data
(2005,1334) ?? 55\4e\4b\4e\4f\57\4e\20 # 8, 1 Unknown Tag & Data
(2005,1335) ?? 55\4e\4b\4e\4f\57\4e\20 # 8, 1 Unknown Tag & Data
(2005,1336) ?? 00\00\00\00 # 4, 1 Unknown Tag & Data
(2005,1337) ?? 00\00\00\00 # 4, 1 Unknown Tag & Data
(2005,1338) ?? 00\00\00\00 # 4, 1 Unknown Tag & Data
(2005,1339) ?? 00\00\00\00 # 4, 1 Unknown Tag & Data
(2005,1340) ?? 50\52\45\5f\46\54 # 6, 1 Unknown Tag & Data
(2005,1341) ?? 55\4e\4b\4e\4f\57\4e\20 # 8, 1 Unknown Tag & Data
(2005,1342) ?? 46\49\44\20 # 4, 1 Unknown Tag & Data
(2005,1343) ?? 59\20 # 2, 1 Unknown Tag & Data
(2005,1344) ?? 00\00\00\00\00\00\00\00\00\00\00\00\00\00\00\00\00\00\00\00\00\00... # 80, 1 Unknown Tag & Data
(2005,1345) ?? 4e\4f # 2, 1 Unknown Tag & Data
(2005,1346) ?? 48\45\52\54\5a\20 # 6, 1 Unknown Tag & Data
(2005,1347) ?? 00\00\00\00 # 4, 1 Unknown Tag & Data
(2005,1348) ?? 4f\46\46\20 # 4, 1 Unknown Tag & Data
(2005,1349) ?? 00\00\00\00 # 4, 1 Unknown Tag & Data
(2005,1350) ?? 00\00\00\00\00\00\00\00 # 8, 1 Unknown Tag & Data
(2005,1351) ?? 00\00 # 2, 1 Unknown Tag & Data
(2005,1352) ?? 00\00 # 2, 1 Unknown Tag & Data
(2005,1355) ?? 00\00\00\00\00\00\00\00\00\00\00\00\00\00\00\00\00\00\00\00\00\00... # 120, 1 Unknown Tag & Data
(2005,1356) ?? 4e\4f # 2, 1 Unknown Tag & Data
(2005,1357) ?? 00\00 # 2, 1 Unknown Tag & Data
(2005,1359) ?? 00\00\80\3f # 4, 1 Unknown Tag & Data
(2005,1360) ?? 00\00\00\00 # 4, 1 Unknown Tag & Data
(2005,1361) ?? 00\00\00\00\00\00\00\00 # 8, 1 Unknown Tag & Data
(2005,1362) ?? 00\00\00\00 # 4, 1 Unknown Tag & Data
(2005,1363) ?? 00\00\00\00 # 4, 1 Unknown Tag & Data
(2005,1364) ?? 4e\4f # 2, 1 Unknown Tag & Data
(2005,1370) ?? 00\00 # 2, 1 Unknown Tag & Data
(2005,1371) ?? (no value available) # 0, 1 Unknown Tag & Data
(2005,1381) ?? 33\20 # 2, 1 Unknown Tag & Data
(2005,1382) ?? 01\00\00\00 # 4, 1 Unknown Tag & Data
(2005,1391) ?? (no value available) # 0, 1 Unknown Tag & Data
(2005,1392) ?? 30\20 # 2, 1 Unknown Tag & Data
(2005,1393) ?? 2d\31 # 2, 1 Unknown Tag & Data
(2005,1396) ?? 4e\4f # 2, 1 Unknown Tag & Data
(2005,1397) ?? 53\4e\4d\33\2e\54\2d\44\31\31\30\30\2e\48\65\61\64\20 # 18, 1 Unknown Tag & Data
(2005,1398) ?? 4e\4f # 2, 1 Unknown Tag & Data
(2005,1399) ?? 4e\4f # 2, 1 Unknown Tag & Data
(2005,1400) ?? 59\45\53\20 # 4, 1 Unknown Tag & Data
(2005,1401) ?? 01\00\00\00 # 4, 1 Unknown Tag & Data
(2005,1402) ?? (no value available) # 0, 1 Unknown Tag & Data
(2005,1403) ?? 00\00\00\00 # 4, 1 Unknown Tag & Data
(2005,1406) ?? 00\00 # 2, 1 Unknown Tag & Data
(2005,1409) ?? 30\20 # 2, 1 Unknown Tag & Data
(2005,140a) ?? 31\20 # 2, 1 Unknown Tag & Data
(2005,140b) ?? 6e\6f\72\6d\61\6c\69\7a\65\64 # 10, 1 Unknown Tag & Data
(2005,140e) ?? (no value available) # 0, 1 Unknown Tag & Data
(2005,140f) SQ (Sequence with undefined length #=1) # u/l, 1 Unknown Tag & Data
(fffe,e000) na (Item with undefined length #=84) # u/l, 1 Item
(0008,0000) UL 102 # 4, 1 IdentifyingGroupLength
(0008,9007) CS [\\\] # 4, 4 FrameType
(0008,9123) UI [1.3.46.670589.11] # 16, 1 CreatorVersionUID
(0008,9205) CS [MONOCHROME] # 10, 1 PixelPresentation
(0008,9206) CS [VOLUME] # 6, 1 VolumetricProperties
(0008,9207) CS [NONE] # 4, 1 VolumeBasedCalculationTechnique
(0008,9208) CS [REAL] # 4, 1 ComplexImageComponent
(0008,9209) CS [T1] # 2, 1 AcquisitionContrast
(0018,0000) UL 844 # 4, 1 AcquisitionGroupLength
(0018,1041) DS [0] # 2, 1 ContrastBolusVolume
(0018,1049) DS [0] # 2, 1 ContrastBolusIngredientConcentration
(0018,9005) SH [T1TFE] # 6, 1 PulseSequenceName
(0018,9008) CS [GRADIENT] # 8, 1 EchoPulseSequence
(0018,9009) CS [NO] # 2, 1 InversionRecovery
(0018,9011) CS [NO] # 2, 1 MultipleSpinEcho
(0018,9012) CS [NO] # 2, 1 MultiPlanarExcitation
(0018,9014) CS [NO] # 2, 1 PhaseContrast
(0018,9015) CS [NO] # 2, 1 TimeOfFlightContrast
(0018,9016) CS [RF] # 2, 1 Spoiling
(0018,9017) CS [LONGITUDINAL] # 12, 1 SteadyStatePulseSequence
(0018,9018) CS [NO] # 2, 1 EchoPlanarPulseSequence
(0018,9020) CS [NONE] # 4, 1 MagnetizationTransfer
(0018,9021) CS [NO] # 2, 1 T2Preparation
(0018,9022) CS [NO] # 2, 1 BloodSignalNulling
(0018,9024) CS [NO] # 2, 1 SaturationRecovery
(0018,9025) CS [FAT] # 4, 1 SpectrallySelectedSuppression
(0018,9026) CS [WATER] # 6, 1 SpectrallySelectedExcitation
(0018,9027) CS [NONE] # 4, 1 SpatialPreSaturation
(0018,9028) CS [NONE] # 4, 1 Tagging
(0018,9029) CS [3D] # 2, 1 OversamplingPhase
(0018,9032) CS [RECTILINEAR] # 12, 1 GeometryOfKSpaceTraversal
(0018,9033) CS [PARTIAL] # 8, 1 SegmentedKSpaceTraversal
(0018,9034) CS [UNKNOWN] # 8, 1 RectilinearPhaseEncodeReordering
(0018,9035) FD 0 # 8, 1 TagThickness
(0018,9037) CS [NONE] # 4, 1 CardiacSynchronizationTechnique
(0018,9041) LO (no value available) # 0, 0 ReceiveCoilManufacturerName
(0018,9043) CS [MULTICOIL] # 10, 1 ReceiveCoilType
(0018,9044) CS [NO] # 2, 1 QuadratureReceiveCoil
(0018,9047) SH [H] # 2, 1 MultiCoilElementName
(0018,9048) CS [YES] # 4, 1 MultiCoilElementUsed
(0018,9051) CS [BODY] # 4, 1 TransmitCoilType
(0018,9053) FD 4.68 # 8, 1 ChemicalShiftReference
(0018,9058) US 100 # 2, 1 MRAcquisitionFrequencyEncodingSteps
(0018,9059) CS [NO] # 2, 1 Decoupling
(0018,9064) CS [RIESZ] # 6, 1 KSpaceFiltering
(0018,9069) FD 1 # 8, 1 ParallelReductionFactorInPlane
(0018,9075) CS [NONE] # 4, 1 DiffusionDirectionality
(0018,9077) CS [YES] # 4, 1 ParallelAcquisition
(0018,9078) CS [SENSE] # 6, 1 ParallelAcquisitionTechnique
(0018,9079) FD 0\0 # 16, 2 InversionTimes
(0018,9080) ST [WATER] # 6, 1 MetaboliteMapDescription
(0018,9081) CS [NO] # 2, 1 PartialFourier
(0018,9090) FD 0\0\0 # 24, 3 VelocityEncodingDirection
(0018,9091) FD 0 # 8, 1 VelocityEncodingMinimumValue
(0018,9093) US 227 # 2, 1 NumberOfKSpaceTrajectories
(0018,9094) CS [ELLIPTICAL] # 10, 1 CoverageOfKSpace
(0018,9101) CS [NO] # 2, 1 FrequencyCorrection
(0018,9147) CS (no value available) # 0, 0 DiffusionAnisotropyType
(0018,9155) FD 1 # 8, 1 ParallelReductionFactorOutOfPlane
(0018,9168) FD 1 # 8, 1 ParallelReductionFactorSecondInPlane
(0018,9170) CS [NONE] # 4, 1 RespiratoryMotionCompensationTechnique
(0018,9171) CS [NONE] # 4, 1 RespiratorySignalSource
(0018,9172) CS [NONE] # 4, 1 BulkMotionCompensationTechnique
(0018,9174) CS [IEC] # 4, 1 ApplicableSafetyStandardAgency
(0018,9179) CS [IEC_WHOLE_BODY] # 14, 1 SpecificAbsorptionRateDefinition
(0018,9180) CS [DB_DT] # 6, 1 GradientOutputType
(0018,9181) FD 0.942994 # 8, 1 SpecificAbsorptionRateValue
(0018,9182) FD 51.1089 # 8, 1 GradientOutput
(0018,9199) CS [NO] # 2, 1 WaterReferencedPhaseCorrection
(0018,9200) CS (no value available) # 0, 0 MRSpectroscopyAcquisitionType
(0018,9231) US 99 # 2, 1 MRAcquisitionPhaseEncodingStepsInPlane
(0018,9232) US 50 # 2, 1 MRAcquisitionPhaseEncodingStepsOutOfPlane
(0018,9240) US 0 # 2, 1 RFEchoTrainLength
(0018,9241) US 30 # 2, 1 GradientEchoTrainLength
(0020,0000) UL 42 # 4, 1 ImageGroupLength
(0020,9072) CS [L] # 2, 1 FrameLaterality
(0020,9254) ?? 00\00\00\00\00\00\00\00 # 8, 1 Unknown Tag & Data
(0020,9255) ?? 00\00\00\00\00\00\00\00 # 8, 1 Unknown Tag & Data
(0028,0000) UL 64 # 4, 1 ImagePresentationGroupLength
(0028,3003) LO [Philips Real World Value Mapping] # 32, 1 LUTExplanation
(0028,9001) UL 1 # 4, 1 DataPointRows
(0028,9002) UL 0 # 4, 1 DataPointColumns
(0040,0000) UL 16 # 4, 1 ModalityWorklistGroupLength
(0040,9210) SH [Philips] # 8, 1 LUTLabel
(fffe,e00d) na (ItemDelimitationItem) # 0, 0 ItemDelimitationItem
(fffe,e0dd) na (SequenceDelimitationItem) # 0, 0 SequenceDelimitationItem
(2005,1412) ?? 31\20 # 2, 1 Unknown Tag & Data
(2005,1413) ?? 31\20 # 2, 1 Unknown Tag & Data
(2005,1414) ?? 01\00\00\00 # 4, 1 Unknown Tag & Data
(2005,1415) ?? 01\00\00\00 # 4, 1 Unknown Tag & Data
(2005,1416) ?? 4d\41\4e\20 # 4, 1 Unknown Tag & Data
(2005,1418) ?? 53\54\41\54\49\43\20\46\49\45\4c\44\5c\52\46\5c\47\52\41\44\49\45... # 24, 1 Unknown Tag & Data
(2005,1419) ?? 49\45\43\5f\46\49\52\53\54\5f\4c\45\56\45\4c\5c\49\45\43\5f\4e\4f... # 38, 1 Unknown Tag & Data
(2005,141a) ?? 53\50\41\49\52\20 # 6, 1 Unknown Tag & Data
(2005,141b) ?? 30\20 # 2, 1 Unknown Tag & Data
(2005,141c) ?? 30\20 # 2, 1 Unknown Tag & Data
(2005,141d) ?? 30\20 # 2, 1 Unknown Tag & Data
(2005,1426) ?? 4e\20 # 2, 1 Unknown Tag & Data
(2005,1428) ?? 00\00\00\00 # 4, 1 Unknown Tag & Data
(2005,142a) ?? 49\4e\49\54\49\41\4c\20 # 8, 1 Unknown Tag & Data
(2005,142b) ?? 49\4e\49\54\49\41\4c\20 # 8, 1 Unknown Tag & Data
(2005,142c) ?? 50\41\52\54\49\41\4c\20 # 8, 1 Unknown Tag & Data
(2005,142d) ?? 49\4e\49\54\49\41\4c\20 # 8, 1 Unknown Tag & Data
(2005,1432) ?? 4e\20 # 2, 1 Unknown Tag & Data
(2005,1435) ?? 4e\20 # 2, 1 Unknown Tag & Data
(2005,143a) ?? 64\61\74\61\64\65\66\73\20\24\52\65\76\69\73\69\6f\6e\3a\20\32\36... # 26, 1 Unknown Tag & Data
(2005,143b) ?? 4e\20 # 2, 1 Unknown Tag & Data
(2005,143f) ?? 4e\20 # 2, 1 Unknown Tag & Data
(2005,1440) ?? 00\00\00\00 # 4, 1 Unknown Tag & Data
(2005,1441) ?? 00\00\00\00 # 4, 1 Unknown Tag & Data
(2005,1442) ?? 00\00\00\00 # 4, 1 Unknown Tag & Data
(2005,1443) ?? 00\00\00\00 # 4, 1 Unknown Tag & Data
(2005,1444) ?? 33\30 # 2, 1 Unknown Tag & Data
(2005,1445) ?? 4e\20 # 2, 1 Unknown Tag & Data
(2005,1447) ?? 00\00\00\00 # 4, 1 Unknown Tag & Data
(2005,144a) ?? 00\00\00\00 # 4, 1 Unknown Tag & Data
(2005,144b) ?? 00\00\00\00 # 4, 1 Unknown Tag & Data
(2005,144c) ?? 00\00\00\00 # 4, 1 Unknown Tag & Data
(2005,144d) ?? 4e\20 # 2, 1 Unknown Tag & Data
(2005,144e) ?? 4e\20 # 2, 1 Unknown Tag & Data
(2005,144f) ?? 4e\4f\4e\45 # 4, 1 Unknown Tag & Data
(2005,1450) ?? 00\00 # 2, 1 Unknown Tag & Data
(2050,0000) UL 16 # 4, 1 PresentationLUTGroupLength
(2050,0020) CS [IDENTITY] # 8, 1 PresentationLUTShape
(7fe0,0000) UL 320008 # 4, 1 PixelDataGroupLength
(7fe0,0010) OW 0575\0575\0575\0575\0575\0575\0575\0575\0575\0575\0575\0575\0575... # 320000, 1 PixelData


Top
 Profile  
 
 Post subject:
PostPosted: Mon, 2011-02-28, 16:54 
Offline
ICSMED DICOM Services
ICSMED DICOM Services

Joined: Fri, 2004-10-29, 21:38
Posts: 2217
Location: Oldenburg, Germany
I guess you are using the "official" Windows binaries from the DCMTK website, right? Then this thread might answer your question ... it's also about the private "Stack Sequence".


Top
 Profile  
 
 Post subject:
PostPosted: Mon, 2011-02-28, 17:52 
Offline

Joined: Tue, 2007-05-08, 17:45
Posts: 183
Yes I'm using the officiel binaries.

I guess I'm dense, but I fail to see where in the dcmdump output a problem is related to tag 2005,105f.

If this is indeed the problem, does that mean that your private tag dictionary changes between 3.5.4 and 3.6.0 are the root cause of this issue?


Top
 Profile  
 
 Post subject:
PostPosted: Mon, 2011-02-28, 18:44 
Offline
ICSMED DICOM Services
ICSMED DICOM Services

Joined: Fri, 2004-10-29, 21:38
Posts: 2217
Location: Oldenburg, Germany
The dataset is stored with LittleEndianImplicit, i.e. without an explicit value representation (VR). The DCMTK, therefore, tries to determine the VR from the data dictionary.
The "official" Windows binaries for DCMTK 3.6.0 contain a built-in version of the private tags data dictionary. And this dictionary contains the following entry ...

Code:
(2005,"Philips MR Imaging DD 001",5f)   SQ   StackSequence   1

... which is apparently not correct for your sample file.

Regarding your question:

Quote:
If this is indeed the problem, does that mean that your private tag dictionary changes between 3.5.4 and 3.6.0 are the root cause of this issue?

Yes, here's the relevant commit.


Top
 Profile  
 
 Post subject:
PostPosted: Tue, 2011-03-01, 09:07 
Offline

Joined: Mon, 2007-09-03, 10:53
Posts: 99
Location: Trondheim, Norway
I am sorry, there is a typo in that contribution. The line

(2005,"Philips MR Imaging DD 001",5f) SQ StackSequence 1

should be changed to

(2001,"Philips MR Imaging DD 001",5f) SQ StackSequence 1

I did not test the change against an implicit data set, and I suppose it looked fine in the dcmdump against my explicit series because it would ignore the above VR anyway.


Top
 Profile  
 
 Post subject:
PostPosted: Tue, 2011-03-01, 11:08 
Offline
ICSMED DICOM Services
ICSMED DICOM Services

Joined: Fri, 2004-10-29, 21:38
Posts: 2217
Location: Oldenburg, Germany
And here is the corresponding commit.


Top
 Profile  
 
 Post subject:
PostPosted: Tue, 2011-03-01, 15:05 
Offline
ICSMED DICOM Services
ICSMED DICOM Services

Joined: Fri, 2004-10-29, 21:38
Posts: 2217
Location: Oldenburg, Germany
We've now also replaced the Windows binaries by a version that does not contain a built-in private dictionary. If required, the "private.dic" can still be loaded by setting the DCMDICTPATH environment variable accordingly.


Top
 Profile  
 
Display posts from previous:  Sort by  
Post new topic Reply to topic  [ 13 posts ] 

All times are UTC + 1 hour


Who is online

Users browsing this forum: Google [Bot] and 1 guest


You cannot post new topics in this forum
You cannot reply to topics in this forum
You cannot edit your posts in this forum
You cannot delete your posts in this forum

Search for:
Jump to:  
Powered by phpBB® Forum Software © phpBB Group