about dcmqrscp
Moderator: Moderator Team
about dcmqrscp
I used dcmqrscp to test 'getscu' which I developped and got an error, the following is the output of dcmqrscp, Is that the role error? could you tell me why? and where should I modified? thanks!
E: Get SCP: storeSCU: [file: /data/db/COMMON/MR_4e2979885f2bbf46.dcm] No presentation context with requestor SCP role for: (MR) 1.2.840.10008.5.1.4.1.1.4
E: getSCP: Get Sub-Op Failed: 0006:0208 DIMSE No valid Presentation Context ID
E: Get SCP: storeSCU: [file: /data/db/COMMON/MR_4e2979885f2bbf46.dcm] No presentation context with requestor SCP role for: (MR) 1.2.840.10008.5.1.4.1.1.4
E: getSCP: Get Sub-Op Failed: 0006:0208 DIMSE No valid Presentation Context ID
Belowing is the output of dcmqrscp(V3.6.0).
D: setting network receive timeout to 60 seconds
W: cannot set network receive timeout to 60 seconds
D: PDU Type: Associate Request, PDU Length: 24826 + 6 bytes PDU header
D: Only dumping 512 bytes.
D: 01 00 00 00 60 fa 00 01 00 00 43 4f 4d 4d 4f 4e
D: 20 20 20 20 20 20 20 20 20 20 4d 45 20 20 20 20
D: 20 20 20 20 20 20 20 20 20 20 00 00 00 00 00 00
D: 00 00 00 00 00 00 00 00 00 00 00 00 00 00 00 00
D: 00 00 00 00 00 00 00 00 00 00 10 00 00 15 31 2e
D: 32 2e 38 34 30 2e 31 30 30 30 38 2e 33 2e 31 2e
D: 31 2e 31 20 00 01 1a 01 00 ff 00 30 00 00 1b 31
D: 2e 32 2e 38 34 30 2e 31 30 30 30 38 2e 35 2e 31
D: 2e 34 2e 31 2e 32 2e 32 2e 33 40 00 00 13 31 2e
D: 32 2e 38 34 30 2e 31 30 30 30 38 2e 31 2e 32 2e
D: 31 40 00 00 11 31 2e 32 2e 38 34 30 2e 31 30 30
D: 30 38 2e 31 2e 32 40 00 00 13 31 2e 32 2e 38 34
D: 30 2e 31 30 30 30 38 2e 31 2e 32 2e 32 40 00 00
D: 16 31 2e 32 2e 38 34 30 2e 31 30 30 30 38 2e 31
D: 2e 32 2e 34 2e 39 30 40 00 00 16 31 2e 32 2e 38
D: 34 30 2e 31 30 30 30 38 2e 31 2e 32 2e 34 2e 39
D: 31 40 00 00 16 31 2e 32 2e 38 34 30 2e 31 30 30
D: 30 38 2e 31 2e 32 2e 34 2e 37 30 40 00 00 16 31
D: 2e 32 2e 38 34 30 2e 31 30 30 30 38 2e 31 2e 32
D: 2e 34 2e 35 30 40 00 00 16 31 2e 32 2e 38 34 30
D: 2e 31 30 30 30 38 2e 31 2e 32 2e 34 2e 35 31 40
D: 00 00 13 31 2e 32 2e 38 34 30 2e 31 30 30 30 38
D: 2e 31 2e 32 2e 35 40 00 00 17 31 2e 32 2e 38 34
D: 30 2e 31 30 30 30 38 2e 31 2e 32 2e 34 2e 31 30
D: 30 20 00 01 1c 03 00 ff 00 30 00 00 1d 31 2e 32
D: 2e 38 34 30 2e 31 30 30 30 38 2e 35 2e 31 2e 34
D: 2e 31 2e 31 2e 39 2e 31 2e 33 40 00 00 13 31 2e
D: 32 2e 38 34 30 2e 31 30 30 30 38 2e 31 2e 32 2e
D: 31 40 00 00 11 31 2e 32 2e 38 34 30 2e 31 30 30
D: 30 38 2e 31 2e 32 40 00 00 13 31 2e 32 2e 38 34
D: 30 2e 31 30 30 30 38 2e 31 2e 32 2e 32 40 00 00
D: 16 31 2e 32 2e 38 34 30 2e 31 30 30 30 38 2e 31
D:
D: Parsing an A-ASSOCIATE PDU
I: Association Received (192.168.1.193:ME -> COMMON) Mon Jul 25 11:04:10 2011
I:
D: Parameters:
D: ====================== BEGIN A-ASSOCIATE-RQ =====================
D: Our Implementation Class UID: 1.2.276.0.7230010.3.0.3.6.0
D: Our Implementation Version Name: OFFIS_DCMTK_360
D: Their Implementation Class UID: 1.2.276.0.7230010.3.0.3.5.4
D: Their Implementation Version Name: OFFIS_DCMTK_354
D: Application Context Name: 1.2.840.10008.3.1.1.1
D: Calling Application Name: ME
D: Called Application Name: COMMON
D: Responding Application Name:
D: Our Max PDU Receive Size: 16384
D: Their Max PDU Receive Size: 16384
D: Presentation Contexts:
D: Context ID: 1 (Proposed)
D: Abstract Syntax: =GETStudyRootQueryRetrieveInformationModel
D: Proposed SCP/SCU Role: Default
D: Proposed Transfer Syntax(es):
D: =LittleEndianExplicit
D: =LittleEndianImplicit
D: =BigEndianExplicit
D: =JPEG2000LosslessOnly
D: =JPEG2000
D: =JPEGLossless:Non-hierarchical-1stOrderPrediction
D: =JPEGBaseline
D: =JPEGExtended:Process2+4
D: =RLELossless
D: =MPEG2MainProfile@MainLevel
D: ...
D: Context ID: 71 (Proposed)
D: Abstract Syntax: =MRImageStorage
D: Proposed SCP/SCU Role: Default
D: Proposed Transfer Syntax(es):
D: =LittleEndianExplicit
D: =LittleEndianImplicit
D: =BigEndianExplicit
D: =JPEG2000LosslessOnly
D: =JPEG2000
D: =JPEGLossless:Non-hierarchical-1stOrderPrediction
D: =JPEGBaseline
D: =JPEGExtended:Process2+4
D: =RLELossless
D: =MPEG2MainProfile@MainLevel
D: Context ID: 73 (Proposed)
D: Abstract Syntax: =MRSpectroscopyStorage
D: Proposed SCP/SCU Role: Default
D: Proposed Transfer Syntax(es):
D: =LittleEndianExplicit
D: =LittleEndianImplicit
D: =BigEndianExplicit
D: =JPEG2000LosslessOnly
D: =JPEG2000
D: =JPEGLossless:Non-hierarchical-1stOrderPrediction
D: =JPEGBaseline
D: =JPEGExtended:Process2+4
D: =RLELossless
D: =MPEG2MainProfile@MainLevel
D: ...
D: Requested Extended Negotiation: none
D: Accepted Extended Negotiation: none
D: Requested User Identity Negotiation: none
D: User Identity Negotiation Response: none
D: ======================= END A-ASSOCIATE-RQ ======================
D: Constructing Associate AC PDU
I: Association Acknowledged (Max Send PDV: 16372)
D: ====================== BEGIN A-ASSOCIATE-AC =====================
D: Our Implementation Class UID: 1.2.276.0.7230010.3.0.3.6.0
D: Our Implementation Version Name: OFFIS_DCMTK_360
D: Their Implementation Class UID: 1.2.276.0.7230010.3.0.3.5.4
D: Their Implementation Version Name: OFFIS_DCMTK_354
D: Application Context Name: 1.2.840.10008.3.1.1.1
D: Calling Application Name: ME
D: Called Application Name: COMMON
D: Responding Application Name:
D: Our Max PDU Receive Size: 16384
D: Their Max PDU Receive Size: 16384
D: Presentation Contexts:
D: Context ID: 1 (Accepted)
D: Abstract Syntax: =GETStudyRootQueryRetrieveInformationModel
D: Proposed SCP/SCU Role: Default
D: Accepted SCP/SCU Role: Default
D: Accepted Transfer Syntax: =LittleEndianExplicit
D: ...
D: Context ID: 71 (Accepted)
D: Abstract Syntax: =MRImageStorage
D: Proposed SCP/SCU Role: Default
D: Accepted SCP/SCU Role: Default
D: Accepted Transfer Syntax: =LittleEndianExplicit
D: Context ID: 73 (Accepted)
D: Abstract Syntax: =MRSpectroscopyStorage
D: Proposed SCP/SCU Role: Default
D: Accepted SCP/SCU Role: Default
D: Accepted Transfer Syntax: =LittleEndianExplicit
D: ...
D: Requested Extended Negotiation: none
D: Accepted Extended Negotiation: none
D: Requested User Identity Negotiation: none
D: User Identity Negotiation Response: none
D: ======================= END A-ASSOCIATE-AC ======================
I: Received Get SCP:
I: ===================== INCOMING DIMSE MESSAGE ====================
I: Message Type : C-GET RQ
I: Message ID : 1
I: Affected SOP Class UID : GETStudyRootQueryRetrieveInformationModel
I: Data Set : present
I: Priority : medium
I: ======================= END DIMSE MESSAGE =======================
I: Get SCP Request Identifiers:
I:
I: # Dicom-Data-Set
I: # Used TransferSyntax: Little Endian Explicit
I: (0008,0052) CS [STUDY] # 6, 1 QueryRetrieveLevel
I: (0020,000d) UI [1.2.840.113619.2.25.1.1762157631.873231325.140] # 46, 1 StudyInstanceUID
I:
E: Get SCP: storeSCU: [file: /data/db/COMMON/MR_4e28f085fbf57199.dcm] No presentation context with requestor SCP role for: (MR) 1.2.840.10008.5.1.4.1.1.4
E: getSCP: Get Sub-Op Failed: 0006:0208 DIMSE No valid Presentation Context ID
I: Get SCP Response 1 [status: Pending]
E: Get SCP: storeSCU: [file: /data/db/COMMON/MR_4e28f54d7ecd5e5a.dcm] No presentation context with requestor SCP role for: (MR) 1.2.840.10008.5.1.4.1.1.4
E: getSCP: Get Sub-Op Failed: 0006:0208 DIMSE No valid Presentation Context ID
I: Get SCP Response 2 [status: Pending]
I: Get SCP Response 3 [status: Refused: OutOfResourcesSubOperations]
I: Association Release
I: Cleaned up after child (1085) Mon Jul 25 11:04:11 2011
I:
D: setting network receive timeout to 60 seconds
W: cannot set network receive timeout to 60 seconds
D: PDU Type: Associate Request, PDU Length: 24826 + 6 bytes PDU header
D: Only dumping 512 bytes.
D: 01 00 00 00 60 fa 00 01 00 00 43 4f 4d 4d 4f 4e
D: 20 20 20 20 20 20 20 20 20 20 4d 45 20 20 20 20
D: 20 20 20 20 20 20 20 20 20 20 00 00 00 00 00 00
D: 00 00 00 00 00 00 00 00 00 00 00 00 00 00 00 00
D: 00 00 00 00 00 00 00 00 00 00 10 00 00 15 31 2e
D: 32 2e 38 34 30 2e 31 30 30 30 38 2e 33 2e 31 2e
D: 31 2e 31 20 00 01 1a 01 00 ff 00 30 00 00 1b 31
D: 2e 32 2e 38 34 30 2e 31 30 30 30 38 2e 35 2e 31
D: 2e 34 2e 31 2e 32 2e 32 2e 33 40 00 00 13 31 2e
D: 32 2e 38 34 30 2e 31 30 30 30 38 2e 31 2e 32 2e
D: 31 40 00 00 11 31 2e 32 2e 38 34 30 2e 31 30 30
D: 30 38 2e 31 2e 32 40 00 00 13 31 2e 32 2e 38 34
D: 30 2e 31 30 30 30 38 2e 31 2e 32 2e 32 40 00 00
D: 16 31 2e 32 2e 38 34 30 2e 31 30 30 30 38 2e 31
D: 2e 32 2e 34 2e 39 30 40 00 00 16 31 2e 32 2e 38
D: 34 30 2e 31 30 30 30 38 2e 31 2e 32 2e 34 2e 39
D: 31 40 00 00 16 31 2e 32 2e 38 34 30 2e 31 30 30
D: 30 38 2e 31 2e 32 2e 34 2e 37 30 40 00 00 16 31
D: 2e 32 2e 38 34 30 2e 31 30 30 30 38 2e 31 2e 32
D: 2e 34 2e 35 30 40 00 00 16 31 2e 32 2e 38 34 30
D: 2e 31 30 30 30 38 2e 31 2e 32 2e 34 2e 35 31 40
D: 00 00 13 31 2e 32 2e 38 34 30 2e 31 30 30 30 38
D: 2e 31 2e 32 2e 35 40 00 00 17 31 2e 32 2e 38 34
D: 30 2e 31 30 30 30 38 2e 31 2e 32 2e 34 2e 31 30
D: 30 20 00 01 1c 03 00 ff 00 30 00 00 1d 31 2e 32
D: 2e 38 34 30 2e 31 30 30 30 38 2e 35 2e 31 2e 34
D: 2e 31 2e 31 2e 39 2e 31 2e 33 40 00 00 13 31 2e
D: 32 2e 38 34 30 2e 31 30 30 30 38 2e 31 2e 32 2e
D: 31 40 00 00 11 31 2e 32 2e 38 34 30 2e 31 30 30
D: 30 38 2e 31 2e 32 40 00 00 13 31 2e 32 2e 38 34
D: 30 2e 31 30 30 30 38 2e 31 2e 32 2e 32 40 00 00
D: 16 31 2e 32 2e 38 34 30 2e 31 30 30 30 38 2e 31
D:
D: Parsing an A-ASSOCIATE PDU
I: Association Received (192.168.1.193:ME -> COMMON) Mon Jul 25 11:04:10 2011
I:
D: Parameters:
D: ====================== BEGIN A-ASSOCIATE-RQ =====================
D: Our Implementation Class UID: 1.2.276.0.7230010.3.0.3.6.0
D: Our Implementation Version Name: OFFIS_DCMTK_360
D: Their Implementation Class UID: 1.2.276.0.7230010.3.0.3.5.4
D: Their Implementation Version Name: OFFIS_DCMTK_354
D: Application Context Name: 1.2.840.10008.3.1.1.1
D: Calling Application Name: ME
D: Called Application Name: COMMON
D: Responding Application Name:
D: Our Max PDU Receive Size: 16384
D: Their Max PDU Receive Size: 16384
D: Presentation Contexts:
D: Context ID: 1 (Proposed)
D: Abstract Syntax: =GETStudyRootQueryRetrieveInformationModel
D: Proposed SCP/SCU Role: Default
D: Proposed Transfer Syntax(es):
D: =LittleEndianExplicit
D: =LittleEndianImplicit
D: =BigEndianExplicit
D: =JPEG2000LosslessOnly
D: =JPEG2000
D: =JPEGLossless:Non-hierarchical-1stOrderPrediction
D: =JPEGBaseline
D: =JPEGExtended:Process2+4
D: =RLELossless
D: =MPEG2MainProfile@MainLevel
D: ...
D: Context ID: 71 (Proposed)
D: Abstract Syntax: =MRImageStorage
D: Proposed SCP/SCU Role: Default
D: Proposed Transfer Syntax(es):
D: =LittleEndianExplicit
D: =LittleEndianImplicit
D: =BigEndianExplicit
D: =JPEG2000LosslessOnly
D: =JPEG2000
D: =JPEGLossless:Non-hierarchical-1stOrderPrediction
D: =JPEGBaseline
D: =JPEGExtended:Process2+4
D: =RLELossless
D: =MPEG2MainProfile@MainLevel
D: Context ID: 73 (Proposed)
D: Abstract Syntax: =MRSpectroscopyStorage
D: Proposed SCP/SCU Role: Default
D: Proposed Transfer Syntax(es):
D: =LittleEndianExplicit
D: =LittleEndianImplicit
D: =BigEndianExplicit
D: =JPEG2000LosslessOnly
D: =JPEG2000
D: =JPEGLossless:Non-hierarchical-1stOrderPrediction
D: =JPEGBaseline
D: =JPEGExtended:Process2+4
D: =RLELossless
D: =MPEG2MainProfile@MainLevel
D: ...
D: Requested Extended Negotiation: none
D: Accepted Extended Negotiation: none
D: Requested User Identity Negotiation: none
D: User Identity Negotiation Response: none
D: ======================= END A-ASSOCIATE-RQ ======================
D: Constructing Associate AC PDU
I: Association Acknowledged (Max Send PDV: 16372)
D: ====================== BEGIN A-ASSOCIATE-AC =====================
D: Our Implementation Class UID: 1.2.276.0.7230010.3.0.3.6.0
D: Our Implementation Version Name: OFFIS_DCMTK_360
D: Their Implementation Class UID: 1.2.276.0.7230010.3.0.3.5.4
D: Their Implementation Version Name: OFFIS_DCMTK_354
D: Application Context Name: 1.2.840.10008.3.1.1.1
D: Calling Application Name: ME
D: Called Application Name: COMMON
D: Responding Application Name:
D: Our Max PDU Receive Size: 16384
D: Their Max PDU Receive Size: 16384
D: Presentation Contexts:
D: Context ID: 1 (Accepted)
D: Abstract Syntax: =GETStudyRootQueryRetrieveInformationModel
D: Proposed SCP/SCU Role: Default
D: Accepted SCP/SCU Role: Default
D: Accepted Transfer Syntax: =LittleEndianExplicit
D: ...
D: Context ID: 71 (Accepted)
D: Abstract Syntax: =MRImageStorage
D: Proposed SCP/SCU Role: Default
D: Accepted SCP/SCU Role: Default
D: Accepted Transfer Syntax: =LittleEndianExplicit
D: Context ID: 73 (Accepted)
D: Abstract Syntax: =MRSpectroscopyStorage
D: Proposed SCP/SCU Role: Default
D: Accepted SCP/SCU Role: Default
D: Accepted Transfer Syntax: =LittleEndianExplicit
D: ...
D: Requested Extended Negotiation: none
D: Accepted Extended Negotiation: none
D: Requested User Identity Negotiation: none
D: User Identity Negotiation Response: none
D: ======================= END A-ASSOCIATE-AC ======================
I: Received Get SCP:
I: ===================== INCOMING DIMSE MESSAGE ====================
I: Message Type : C-GET RQ
I: Message ID : 1
I: Affected SOP Class UID : GETStudyRootQueryRetrieveInformationModel
I: Data Set : present
I: Priority : medium
I: ======================= END DIMSE MESSAGE =======================
I: Get SCP Request Identifiers:
I:
I: # Dicom-Data-Set
I: # Used TransferSyntax: Little Endian Explicit
I: (0008,0052) CS [STUDY] # 6, 1 QueryRetrieveLevel
I: (0020,000d) UI [1.2.840.113619.2.25.1.1762157631.873231325.140] # 46, 1 StudyInstanceUID
I:
E: Get SCP: storeSCU: [file: /data/db/COMMON/MR_4e28f085fbf57199.dcm] No presentation context with requestor SCP role for: (MR) 1.2.840.10008.5.1.4.1.1.4
E: getSCP: Get Sub-Op Failed: 0006:0208 DIMSE No valid Presentation Context ID
I: Get SCP Response 1 [status: Pending]
E: Get SCP: storeSCU: [file: /data/db/COMMON/MR_4e28f54d7ecd5e5a.dcm] No presentation context with requestor SCP role for: (MR) 1.2.840.10008.5.1.4.1.1.4
E: getSCP: Get Sub-Op Failed: 0006:0208 DIMSE No valid Presentation Context ID
I: Get SCP Response 2 [status: Pending]
I: Get SCP Response 3 [status: Refused: OutOfResourcesSubOperations]
I: Association Release
I: Cleaned up after child (1085) Mon Jul 25 11:04:11 2011
I:
-
- Posts: 101
- Joined: Wed, 2009-07-08, 16:06
- Location: Oldenburg, Germany
The default role means that the association requestor is the SCU and the association acceptor is the SCP. For a successful C-Get operation, the c-store operations must reverse the roles. So, your getscu should propose SCP role for MR image storage. If you are using the last snapshot, you will find the function addPresentationContext() in DCMSCU class that takes the role of the SOP class as a parameter.
-
- Posts: 101
- Joined: Wed, 2009-07-08, 16:06
- Location: Oldenburg, Germany
Thank you for your help!
I modifed getscu.cc, like the following, still not work, could you give me some further hints ?
/*
** Network initialization structure.
** Is a wrapper for DUL functionality.
T_ASC_NetworkRole role = NET_ACCEPTORREQUESTOR;//(opt_retrievePort > 0) ? NET_ACCEPTORREQUESTOR : NET_REQUESTOR;
OFCondition cond = ASC_initializeNetwork(role, OFstatic_cast(int, opt_retrievePort), opt_acse_timeout, &net);
I modifed getscu.cc, like the following, still not work, could you give me some further hints ?
/*
** Network initialization structure.
** Is a wrapper for DUL functionality.
T_ASC_NetworkRole role = NET_ACCEPTORREQUESTOR;//(opt_retrievePort > 0) ? NET_ACCEPTORREQUESTOR : NET_REQUESTOR;
OFCondition cond = ASC_initializeNetwork(role, OFstatic_cast(int, opt_retrievePort), opt_acse_timeout, &net);
-
- Posts: 101
- Joined: Wed, 2009-07-08, 16:06
- Location: Oldenburg, Germany
This is not the correct parameter place to select SCP role for association negotiation. Role selection is specified during association negotiation not in the initialization of the network. You need to add it to the function ASC_addPresentationContext
By default the role is SCU for the association requestor. But for C-GET, your proposed role in this case should be ASC_SC_ROLE_SCP for each of the storage SOP classes you wish to get.OFCondition
ASC_addPresentationContext(
T_ASC_Parameters * params,
T_ASC_PresentationContextID presentationContextID,
const char* abstractSyntax,
const char* transferSyntaxList[],
int transferSyntaxListCount,
T_ASC_SC_ROLE ´= ASC_SC_ROLE_DEFAULT);
-
- Posts: 101
- Joined: Wed, 2009-07-08, 16:06
- Location: Oldenburg, Germany
-
- Posts: 28
- Joined: Tue, 2011-06-28, 13:36
- Location: Chennai
- Contact:
C-GET
Can u please send the code to this mail id.(manoindia2020@gmail.com) I am seriously searching for a source to implement C-GET method.Xiao wrote:I have modified dimget.cc, and generate getscu.cc, if you would like the modification to be integrated into the dcmtk package, I will send you the code.(too big to post it here)
with regards
Manoj Kumar D
Manoj Kumar D
-
- DCMTK Developer
- Posts: 2049
- Joined: Fri, 2004-11-05, 13:47
- Location: Oldenburg, Germany
- Contact:
-
- DCMTK Developer
- Posts: 2049
- Joined: Fri, 2004-11-05, 13:47
- Location: Oldenburg, Germany
- Contact:
Hi,
everyhting you need for doing C-GET as a client including a ready-to-use tool is now released, see DCMTK's public git repository. This is the main commit followed by some enhancements made by Jörg.
Best regards,
Michael
everyhting you need for doing C-GET as a client including a ready-to-use tool is now released, see DCMTK's public git repository. This is the main commit followed by some enhancements made by Jörg.
Best regards,
Michael
-
- Posts: 28
- Joined: Tue, 2011-06-28, 13:36
- Location: Chennai
- Contact:
getscu.cc sample command-line arguments
Hi,
I have integrated the getscu in my application and tried to call the getscu using
getscu 11112 www.dicomserver.co.uk Querydicom.dcm
but the app crashes. I will try to fix by my side. Meanwhile Can I get sample arguments and how the files will be saved in disk?
I have integrated the getscu in my application and tried to call the getscu using
getscu 11112 www.dicomserver.co.uk Querydicom.dcm
but the app crashes. I will try to fix by my side. Meanwhile Can I get sample arguments and how the files will be saved in disk?
with regards
Manoj Kumar D
Manoj Kumar D
-
- DCMTK Developer
- Posts: 2049
- Joined: Fri, 2004-11-05, 13:47
- Location: Oldenburg, Germany
- Contact:
Hi,
It crashes? Thats bad Sure it's not your app? What is your exact call and what is inside your query file?
Best regards,
Michael
Edit: Works for me:
cgetscu dicomserver.co.uk 104 -P query.dcm -v
I: Requesting Association
I: Association Accepted (Max Send PDV: 65524)
I: Send C-GET Request
I: Received C-STORE Request
I: Sending C-STORE Response
I: Received C-GET Response
I: Received C-STORE Request
I: Sending C-STORE Response
I: Received C-GET Response
I: Received C-STORE Request
I: Sending C-STORE Response
I: Received C-GET Response
I: Received C-STORE Request
I: Sending C-STORE Response
I: Received C-GET Response
I: Received C-STORE Request
I: Sending C-STORE Response
I: Received C-GET Response
I: Received C-STORE Request
I: Sending C-STORE Response
I: Received C-GET Response
I: Received C-GET Response
I: Final status reports from last C-GET message:
I: Number of Remaining Suboperations: 0
I: Number of Completed Suboperations: 6
I: Number of Failed Suboperations : 0
I: Number of Warning Suboperations : 0
I: Releasing Association
I: Releasing Association
with query.dcm consisting of
# Dicom-Data-Set
# Used TransferSyntax: Little Endian Explicit
(0008,0052) CS [PATIENT] # 8, 1 QueryRetrieveLevel
(0010,0020) LO [fff] # 4, 1 PatientID
It crashes? Thats bad Sure it's not your app? What is your exact call and what is inside your query file?
Best regards,
Michael
Edit: Works for me:
cgetscu dicomserver.co.uk 104 -P query.dcm -v
I: Requesting Association
I: Association Accepted (Max Send PDV: 65524)
I: Send C-GET Request
I: Received C-STORE Request
I: Sending C-STORE Response
I: Received C-GET Response
I: Received C-STORE Request
I: Sending C-STORE Response
I: Received C-GET Response
I: Received C-STORE Request
I: Sending C-STORE Response
I: Received C-GET Response
I: Received C-STORE Request
I: Sending C-STORE Response
I: Received C-GET Response
I: Received C-STORE Request
I: Sending C-STORE Response
I: Received C-GET Response
I: Received C-STORE Request
I: Sending C-STORE Response
I: Received C-GET Response
I: Received C-GET Response
I: Final status reports from last C-GET message:
I: Number of Remaining Suboperations: 0
I: Number of Completed Suboperations: 6
I: Number of Failed Suboperations : 0
I: Number of Warning Suboperations : 0
I: Releasing Association
I: Releasing Association
with query.dcm consisting of
# Dicom-Data-Set
# Used TransferSyntax: Little Endian Explicit
(0008,0052) CS [PATIENT] # 8, 1 QueryRetrieveLevel
(0010,0020) LO [fff] # 4, 1 PatientID