According to the man page Storescp is conformant to the Procedural Event Logging SOP.
Code: Select all
ProcedureLogStorage 1.2.840.10008.5.1.4.1.1.88.40
As described in
FAQ#35 conformance statements are not provided for DCMTK.
How can storescp be notified of the N-Action procedural event logging as described in P.2.5.2 SCP Conformance (
http://dicom.nema.org/dicom/2013/output ... t_P.2.html )?
If an SCP can be configured to support procedural event logging, the premise is availability monitoring of medical appliances could be possible.
J. Riesmeier
DCMTK Developer
Posts: 2506 Joined: Tue, 2011-05-03, 14:38
Location: Oldenburg, Germany
Contact:
#2
Post
by J. Riesmeier » Mon, 2017-05-01, 10:36
According to the man page Storescp is conformant to the Procedural Event Logging SOP.
Code: Select all
ProcedureLogStorage 1.2.840.10008.5.1.4.1.1.88.40
No, the storescp supports the Procedure Log Storage SOP Class, i.e. you can send instances of this class using C-STORE, but not the Procedural Event Logging SOP Class, which is a totally different thing.
Btw, you should also use the latest edition of the DICOM standard, which is currently "2017a":
http://dicom.nema.org/medical/dicom/current/