I have a question about using wildcards in patient id query of modality worklist.
It seems that I can query patient name, accession number, requested procedure ID, using wildcards and retrieve matching records, eg:
patient name = D* will return all patients starting with the letter D
-but-
the same doesn't apply to ID?
Looking at IHE and other efforts, it seems like it is a requirement to be
able to query on patient ID wildcards as well.
Should this be fixed in wlfsim, or was it done that way for another reason, and would I be breaking something by making this fix?
wlfsim query patient id wildcards
Moderator: Moderator Team
-
- Posts: 25
- Joined: Wed, 2005-07-20, 22:42
- Location: Payson, Arizona
-
- Posts: 25
- Joined: Wed, 2005-07-20, 22:42
- Location: Payson, Arizona
To be specific, I would propose this change in wlfsim.cxx:
[code]
// ----------------------------------------------------------------------------
OFBool WlmFileSystemInteractionManager::PatientIdsMatch( const char *datasetValue, const char *searchMaskValue )
// Date : July 12, 2002
// Author : Thomas Wilkens
// Task : This function returns OFTrue if the dataset's and the search mask's values in
// attribute patient id match; otherwise OFFalse will be returned.
// Parameters : datasetValue - [in] Value for the corresponding attribute in the dataset; might be NULL.
// searchMaskValue - [in] Value for the corresponding attribute in the search mask; never NULL.
// Return Value : OFTrue if the values match, OFFalse otherwise.
{
// if there is a value in the dataset, perform case sensitive single value matching
if( datasetValue != NULL )
// roy 04-11-2006 add wildcards... ===>
return( CaseSensitiveSingleValueMatch( datasetValue, searchMaskValue ) || WildcardMatch( datasetValue, searchMaskValue ) );
else
{
// if datasetValue is not existent, the search mask's value has to be empty to have
// a match (universal matching); in all other cases, we do not have a match
if( strcmp( searchMaskValue, "" ) == 0 )
return( OFTrue );
else
return( OFFalse );
}
}
[/code]
[code]
// ----------------------------------------------------------------------------
OFBool WlmFileSystemInteractionManager::PatientIdsMatch( const char *datasetValue, const char *searchMaskValue )
// Date : July 12, 2002
// Author : Thomas Wilkens
// Task : This function returns OFTrue if the dataset's and the search mask's values in
// attribute patient id match; otherwise OFFalse will be returned.
// Parameters : datasetValue - [in] Value for the corresponding attribute in the dataset; might be NULL.
// searchMaskValue - [in] Value for the corresponding attribute in the search mask; never NULL.
// Return Value : OFTrue if the values match, OFFalse otherwise.
{
// if there is a value in the dataset, perform case sensitive single value matching
if( datasetValue != NULL )
// roy 04-11-2006 add wildcards... ===>
return( CaseSensitiveSingleValueMatch( datasetValue, searchMaskValue ) || WildcardMatch( datasetValue, searchMaskValue ) );
else
{
// if datasetValue is not existent, the search mask's value has to be empty to have
// a match (universal matching); in all other cases, we do not have a match
if( strcmp( searchMaskValue, "" ) == 0 )
return( OFTrue );
else
return( OFFalse );
}
}
[/code]
-
- OFFIS DICOM Team
- Posts: 1445
- Joined: Tue, 2004-11-02, 17:22
- Location: Oldenburg, Germany
- Contact:
I'd be interested to know where in the IHE Technical Framework you derive this requirement from. In fact, the DICOM standard explicitly forbids wildcard matching on the Patient ID, so I would be very surprised if IHE should mandate it. DICOM Part 4, Table K.6-1 specifies that "Patient ID shall be retrieved with Single Value Matching."Looking at IHE and other efforts, it seems like it is a requirement to be able to query on patient ID wildcards as well.
-
- Posts: 25
- Joined: Wed, 2005-07-20, 22:42
- Location: Payson, Arizona
Thanks, I must have been misunderstanding what I was referring to in the IHE tech framework...
The specific document I was looking at in this case was:
IHE Technical Framework vol II Transactions, rev 6.0, May 18, 2005
and within this,
4.5 Query Modality Worklist
then in section
4.5.4.1.2.1 Examples for the Use of Matching Key Attributes
[Quote]
Note: Applications are recommended to append a wildcard "*", if one was not previously entered by the user, at the end of each component of the structured Patient Name.
[/Quote]
Looks like I was confused by this to mean extending wildcards to all components of the query.
But that is precisely why I posted this question, rather than simply implementing the code as outlined, which would have been incorrect!
Many thanks for the response
The specific document I was looking at in this case was:
IHE Technical Framework vol II Transactions, rev 6.0, May 18, 2005
and within this,
4.5 Query Modality Worklist
then in section
4.5.4.1.2.1 Examples for the Use of Matching Key Attributes
[Quote]
Note: Applications are recommended to append a wildcard "*", if one was not previously entered by the user, at the end of each component of the structured Patient Name.
[/Quote]
Looks like I was confused by this to mean extending wildcards to all components of the query.
But that is precisely why I posted this question, rather than simply implementing the code as outlined, which would have been incorrect!
Many thanks for the response
Who is online
Users browsing this forum: No registered users and 1 guest