storescp problems

All other questions regarding DCMTK

Moderator: Moderator Team

Post Reply
Message
Author
santosh
Posts: 61
Joined: Tue, 2007-07-31, 09:27
Location: Bangalore (India)

storescp problems

#1 Post by santosh »

Hi All,

I am using the following command to store the images coming from the GE ct modality model name ProSpeed,

storescp -v -d -aet radServer 5001

now the images which i receive have small white squares with black background, i ran the dicom scp provided by dicom eye and the images are getting stored properly.
The dicom header information is as follows,

0002,0000,File Meta Elements Group Len=194
0002,0001,File Meta Info Version=256
0002,0002,Media Storage SOP Class UID=1.2.840.10008.5.1.4.1.1.2.
0002,0003,Media Storage SOP Inst UID=1.2.840.113619.2.3.281.6006.2008.12.9.14117.1.1.1.
0002,0010,Transfer Syntax UID=1.2.840.10008.1.2.2.
0002,0012,Implementation Class UID=1.2.276.0.7230010.3.0.3.5.4.
0002,0013,Implementation Version Name=OFFIS.DCMTK.354
0008,0008,Image Type=LOCALIZER
0008,0016,SOP Class UID=1.2.840.10008.5.1.4.1.1.2.
0008,0018,SOP Instance UID=1.2.840.113619.2.3.281.6006.2008.12.9.14117.1.1.1.
0008,0020,Study Date=20081209
0008,0022,Acquisition Date=20081209
0008,0023,Image Date=
0008,0030,Study Time=
0008,0032,Acquisition Time=
0008,0033,Image Time=
0008,0050,Accession Number=
0008,0060,Modality=CT
0008,0070,Manufacturer=GE MEDICAL SYSTEMS
0008,0080,Institution Name=
0008,0090,Referring Physician's Name=
0008,1010,Station Name=MKS-CT1
0008,1030,Study Description=NCCT HEAD
0008,103E,Series Description=NCCT HEAD
0008,1090,Manufacturer's Model Name=ProSpeed
0009,0010,?=GEMS.IDEN.01
0009,1004,?=ProSpeed
0009,101A,?=3
0009,10E2,?=ORIGINAL\PROSPECTIVE\LOCALIZER
0009,10E3,?=1.2.840.113619.1.3.281.6006.
0010,0010,Patient's Name=
0010,0020,Patient ID=
0010,0030,Patient Date of Birth=
0010,0040,Patient Sex=M
0010,1010,Patient Age=
0018,0010,Contrast/Bolus Agent=NONE
0018,0015,Body Part Examined=1212498244
0018,0022,Scan Options=SCOUT MODE
0018,0050,Slice Thickness=
0018,0060,KVP [Peak KV]=120.0
0018,1000,Device Serial Number=01196006
0018,1020,Software Version=0604
0018,1120,Gantry/Detector Tilt=808333344
0018,1130,Table Height=143.000000
0018,1140,Rotation Direction=17239
0018,1150,Exposure Time [ms]=6133
0018,1151,X-ray Tube Current [mA]=60
0018,1152,Acquisition Device Processing Description=859191328
0018,1200,Date of Last Calibration=20081209
0018,1201,Time of Last Calibration=005327.000
0018,1210,Convolution Kernel=STANDARD
0018,5100,Patient Position=HFS
0019,0010,?=GEMS.ACQU.01
0019,1025,?=0
0019,1026,?=90
0019,103E,?=0
0019,103F,?=0
0019,1042,?=0
0019,1043,?=0
0019,10F8,?=0
0020,000D,Study Instance UID=1.2.840.113619.2.3.281.6006.2008.12.9.14117.
0020,000E,Series Instance UID=1.2.840.113619.2.3.281.6006.2008.12.9.14117.1.
0020,0010,Study ID=14117
0020,0011,Series Number=1
0020,0012,Acquisition Number=
0020,0013,Image Number=1
0020,0032,Image Position Patient=0.000000\-127.750000\183.000000
0020,0037,Image Orientation (Patient)=0.000000\0.000000\-1.000000\0.000000\1.000000\0.000000
0020,0052,Frame of Reference UID=1.2.840.113619.2.3.281.6006.2008.12.9.14117.20301.
0020,0060,Laterality=
0020,1040,Position Reference=OM
0021,0010,?=GEMS.RELA.01
0021,1070,?=1.2.840.113619.2.3.281.6006.2008.12.9.14117.
0021,1071,?=1.2.840.113619.2.3.281.6006.2008.12.9.14117.1.
0028,0002,Samples Per Pixel=1
0028,0004,Photometric Interpretation=MONOCHROME2
0028,0010,Rows=512
0028,0011,Columns=512
0028,0030,Pixel Spacing=0.500000 0.500000
0028,0100,Bits Allocated=16
0028,0101,Bits Stored=16
0028,0102,High Bit=15
0028,0103,Pixel Representation=1
0028,0106,Smallest Image Pixel Value=57344
0028,0107,Largest Image Pixel Value=8191
0028,0120,Pixel Padding Value=57344
0028,1050,Window Center=110
0028,1051,Window Width=300
0028,1052,Rescale Intercept=0
0028,1053,Rescale Slope=1
0033,0010,?=GEMS.YMHD.01
0033,1005,?=.STD.....v14117 2008/12/0904.58.24.43814117 35. M ..YOKOGAWA MEDICAL SYSTEMSProSpeed ............NO C.................K..o.......SYS ..DATAFILEIMAGEINFVR............NCCT HEAD ......
0033,1006,?=.IMG.....v14117 010101..........2008/12/0904.59.51.098..........H SPHEAD OM SAG B.......................25.0 1............SCLT 6.1 SC ......NO C ............ 90 +0.0 ...p..C...+00.0... 6133 60 368 120 CW S.............v14117.010101.................................................K.0.X25-CAL H A F P ....................................P STND ........25.0.....................n...................... ............................................................................SYS ..DATAFILEIMAGE VR....NCCT HEAD
7FE0,0010,Pixel Data=524288


Please let me know what is going wrong here.
Thanks in advance
Santosh S B

Jörg Riesmeier
ICSMED DICOM Services
ICSMED DICOM Services
Posts: 2217
Joined: Fri, 2004-10-29, 21:38
Location: Oldenburg, Germany

#2 Post by Jörg Riesmeier »

As far as I know, DICOM Eye uses the same code basis for the network functionality as DCMTK's storescp.

So, if I understand you correctly the image is displayed differently on the same DICOM viewer if you received it by storescp or DICOM Eye?

santosh
Posts: 61
Joined: Tue, 2007-07-31, 09:27
Location: Bangalore (India)

#3 Post by santosh »

Hi Jorg,

I used the ezDicom to view the images. The images received by Dicom Eye are displayed properly were as those received by storescp are black and white.
Santosh S B

Jörg Riesmeier
ICSMED DICOM Services
ICSMED DICOM Services
Posts: 2217
Joined: Fri, 2004-10-29, 21:38
Location: Oldenburg, Germany

#4 Post by Jörg Riesmeier »

Before we start guessing what the reason might be, could you please send both images (the one received by storescp and the one received by DICOM Eye) to us by email: dicom/at/offis/dot/de.

santosh
Posts: 61
Joined: Tue, 2007-07-31, 09:27
Location: Bangalore (India)

#5 Post by santosh »

Dear Jorg,

I have mailed both the dicom files requested, please go through them and let me know the findings.
Santosh S B

santosh
Posts: 61
Joined: Tue, 2007-07-31, 09:27
Location: Bangalore (India)

#6 Post by santosh »

i used the --implicit option and the images are perfectly fine..
Thanks for the support
Santosh S B

Post Reply