Hello all,
I am working with MR modality images. Sometimes these images have tags rescale intercept/rescale slope. Reading comp.protocols.dicom the posts say that you should not use intercept/slope on MR modality images.
So does the class dcmimgle/dcmimage which uses intercept/slope before giving you the pixel data take this into account if modailty is MR? Or can MR still use (and should?) really use these tags too?
Thanks,
Mark
slope/intercept in dcmimgle/dcmimage
Moderator: Moderator Team
-
- ICSMED DICOM Services
- Posts: 2217
- Joined: Fri, 2004-10-29, 21:38
- Location: Oldenburg, Germany
The module dcmimgle always [1] uses the values of Rescale Slope/Intercept if these attributes are present in the dataset. If you don't want this you have the choice to disable the modality LUT transformation by setting the flag CIF_IgnoreModalityTransformation in the constructor of class DicomImage.
[1] Apart from XA and XRF images as required by the DICOM standard.
[1] Apart from XA and XRF images as required by the DICOM standard.