Hello i'm using storeUser to send an image to PAC Server but it's strange because i don't tell him to store at any patient where he storage the image?

And if i send an image that already exists it works fine but if i send an image that there's no connection with any patient at PAC Server it crashes saying
'DIMSE failed to send message'

martinrame
Posts: 329
Joined: Mon, 2009-02-23, 19:57

#2 Post by martinrame » Mon, 2011-05-23, 23:52

Forward, could you post the entire command you are using to send the image to the PACS?

Also, please paste the results of the storescu command with -d (debug) parameter.

Forward
Posts: 17
Joined: Mon, 2011-05-02, 11:54

#3 Post by Forward » Tue, 2011-05-24, 08:51

Hey, i'm not calling storescu.exe directly i use c++

After opening the connection i do this to send image

Code: Select all

 retcode = dcmff.loadFile(filename, EXS_Unknown, EGL_noChange, DCM_MaxReadLength, ERM_autoDetect);
   VERIFY_ERROR(retcode);

   if (!DU_findSOPClassAndInstanceInDataSet(dcmff.getDataset(), sopClass, sopInstance)) {
      return false;
   }
  
   DcmXfer filexfer(dcmff.getDataset()->getOriginalXfer());
   if (filexfer.isNotEncapsulated())
      filexfer = EXS_DeflatedLittleEndianExplicit;
   
   if (filexfer.getXfer() != EXS_Unknown)
      presID = ASC_findAcceptedPresentationContextID(m_assoc, sopClass, filexfer.getXferID());
   else
      presID = ASC_findAcceptedPresentationContextID(m_assoc, sopClass);
    
   req.MessageID   = m_assoc->nextMsgID++;
   req.DataSetType = DIMSE_DATASET_PRESENT;
   req.Priority    = DIMSE_PRIORITY_MEDIUM;   
   strcpy(req.AffectedSOPClassUID, sopClass);
   strcpy(req.AffectedSOPInstanceUID, sopInstance);
   strcpy(req.MoveOriginatorApplicationEntityTitle, m_scu_aetitle);
  
   retcode = DIMSE_storeUser(m_assoc, presID, &req, (const char *)filename, dcmff.getDataset(), NULL, NULL, DIMSE_BLOCKING, 5, &rsp, &statusDetail, NULL, OFStandard::getFileSize((const char *)filename));
   VERIFY_ERROR(retcode);
return true;
Hope it helps

Forward
Posts: 17
Joined: Mon, 2011-05-02, 11:54

#4 Post by Forward » Wed, 2011-05-25, 17:31

I tried to send the image now with storescu and the message was...
Unable to convert dataset from 'JPEG 2000 (Lossless or Lossy)' transfer syntax to ' Little Endian Implicit'

Michael Onken
DCMTK Developer
Posts: 1815
Joined: Fri, 2004-11-05, 13:47
Location: Oldenburg, Germany
Contact:

#5 Post by Michael Onken » Thu, 2011-05-26, 09:28

Hi,

this is because storescu does not automatically compress/decompress any images and usually it only negotiates uncompressed transfer syntaxes for image transfer with the server.

Try using option -xv and -xw and see whether it works. This will also let storescu propose JPEG2000 transfer syntaxes,i.e. avoid the decompression.

Best regards,
Michael

Forward
Posts: 17
Joined: Mon, 2011-05-02, 11:54

#6 Post by Forward » Thu, 2011-05-26, 17:10

Hmm tried, but didn't worked

Michael Onken
DCMTK Developer
Posts: 1815
Joined: Fri, 2004-11-05, 13:47
Location: Oldenburg, Germany
Contact:

#7 Post by Michael Onken » Thu, 2011-05-26, 17:27

Hi,

first: Run dcmdump on your file and tell me what it says in the attribute value for "SOP Class UID".

Then run storescu with -d and tell me the output.

Michael

Forward
Posts: 17
Joined: Mon, 2011-05-02, 11:54

#8 Post by Forward » Thu, 2011-05-26, 17:44

dcmdump

dcmdump.exe c:\ll.dcm

Code: Select all

# Dicom-File-Format

# Dicom-Meta-Information-Header
# Used TransferSyntax: Little Endian Explicit
(0002,0000) UL 190                                      #   4, 1 FileMetaInforma
tionGroupLength
(0002,0001) OB 00\01                                    #   2, 1 FileMetaInforma
tionVersion
(0002,0002) UI =CTImageStorage                          #  26, 1 MediaStorageSOP
ClassUID
(0002,0003) UI [1.2.840.113704.1.111.3352.1107859679.1269] #  42, 1 MediaStorage
SOPInstanceUID
(0002,0010) UI =JPEG2000                                #  22, 1 TransferSyntaxU
ID
(0002,0012) UI [1.3.6.1.4.1.19291.2.1]                  #  22, 1 ImplementationC
lassUID
(0002,0013) SH [OSIRIX001]                              #  10, 1 ImplementationV
ersionName
(0002,0016) AE [OsiriX]                                 #   6, 1 SourceApplicati
onEntityTitle

# Dicom-Data-Set
# Used TransferSyntax: JPEG 2000 (Lossless or Lossy)
(0008,0005) CS [ISO_IR 100]                             #  10, 1 SpecificCharact
erSet
(0008,0008) CS [ORIGINAL\PRIMARY\AXIAL\HELIX]           #  28, 4 ImageType
(0008,0012) DA [20050201]                               #   8, 1 InstanceCreatio
nDate
(0008,0013) TM [120000.000000]                          #  14, 1 InstanceCreatio
nTime
(0008,0016) UI =CTImageStorage                          #  26, 1 SOPClassUID
(0008,0018) UI [1.2.840.113704.1.111.3352.1107859679.1269] #  42, 1 SOPInstanceU
ID
(0008,0020) DA [20050201]                               #   8, 1 StudyDate
(0008,0022) DA [20050208]                               #   8, 1 AcquisitionDate

(0008,0023) DA [20050208]                               #   8, 1 ContentDate
(0008,0030) TM [120000.000000]                          #  14, 1 StudyTime
(0008,0032) TM [114308.000000]                          #  14, 1 AcquisitionTime

(0008,0033) TM [114308.907000]                          #  14, 1 ContentTime
(0008,0050) SH [1963943]                                #   8, 1 AccessionNumber

(0008,0060) CS [CT]                                     #   2, 1 Modality
(0008,0070) LO [Philips]                                #   8, 1 Manufacturer
(0008,0080) LO [dPJ]                                    #   4, 1 InstitutionName

(0008,0081) ST [GENEVA]                                 #   6, 1 InstitutionAddr
ess
(0008,0090) PN [NULL^DUMCOB]                            #  12, 1 ReferringPhysic
ianName
(0008,1010) SH [ctdumc]                                 #   6, 1 StationName
(0008,1030) LO [CER/CT]                                 #   6, 1 StudyDescriptio
n
(0008,1032) SQ (Sequence with explicit length #=1)      #  36, 1 ProcedureCodeSe
quence
  (fffe,e000) na (Item with explicit length #=2)          #  28, 1 Item
    (0008,0102) SH [BROKER]                                 #   6, 1 CodingSchem
eDesignator
    (0008,0104) LO [CER/CT]                                 #   6, 1 CodeMeaning

  (fffe,e00d) na (ItemDelimitationItem for re-encoding)   #   0, 0 ItemDelimitat
ionItem
(fffe,e0dd) na (SequenceDelimitationItem for re-encod.) #   0, 0 SequenceDelimit
ationItem
(0008,103e) LO [ANGIO CT]                               #   8, 1 SeriesDescripti
on
(0008,1040) LO [CAU]                                    #   4, 1 InstitutionalDe
partmentName
(0008,1070) PN [CTDUMC\mxservice]                       #  16, 2 OperatorsName
(0008,1090) LO [Mx8000 IDT]                             #  10, 1 ManufacturerMod
elName
(0008,1110) SQ (Sequence with explicit length #=1)      # 100, 1 ReferencedStudy
Sequence
  (fffe,e000) na (Item with explicit length #=2)          #  92, 1 Item
    (0008,1150) UI =RETIRED_DetachedStudyManagementSOPClass #  24, 1 ReferencedS
OPClassUID
    (0008,1155) UI [1.2.124.113532.129.195.3.60.20050208.112149.1708074] #  52,
1 ReferencedSOPInstanceUID
  (fffe,e00d) na (ItemDelimitationItem for re-encoding)   #   0, 0 ItemDelimitat
ionItem
(fffe,e0dd) na (SequenceDelimitationItem for re-encod.) #   0, 0 SequenceDelimit
ationItem
(0008,1120) SQ (Sequence with explicit length #=1)      # 100, 1 ReferencedPatie
ntSequence
  (fffe,e000) na (Item with explicit length #=2)          #  92, 1 Item
    (0008,1150) UI =RETIRED_DetachedPatientManagementSOPClass #  24, 1 Reference
dSOPClassUID
    (0008,1155) UI [1.2.124.113532.129.195.3.60.20030224.174643.1093514] #  52,
1 ReferencedSOPInstanceUID
  (fffe,e00d) na (ItemDelimitationItem for re-encoding)   #   0, 0 ItemDelimitat
ionItem
(fffe,e0dd) na (SequenceDelimitationItem for re-encod.) #   0, 0 SequenceDelimit
ationItem
(0008,1140) SQ (Sequence with explicit length #=1)      #  92, 1 ReferencedImage
Sequence
  (fffe,e000) na (Item with explicit length #=2)          #  84, 1 Item
    (0008,1150) UI =CTImageStorage                          #  26, 1 ReferencedS
OPClassUID
    (0008,1155) UI [1.2.840.113704.1.111.3352.1107859195.1185] #  42, 1 Referenc
edSOPInstanceUID
  (fffe,e00d) na (ItemDelimitationItem for re-encoding)   #   0, 0 ItemDelimitat
ionItem
(fffe,e0dd) na (SequenceDelimitationItem for re-encod.) #   0, 0 SequenceDelimit
ationItem
(0010,0010) PN [MANIX]                                  #   6, 1 PatientName
(0010,0020) LO [7Thjq7g]                                #   8, 1 PatientID
(0010,1010) AS [000Y]                                   #   4, 1 PatientAge
(0018,0010) LO [CONTRAST]                               #   8, 1 ContrastBolusAg
ent
(0018,0022) CS [HELIX]                                  #   6, 1 ScanOptions
(0018,0050) DS [1.5]                                    #   4, 1 SliceThickness
(0018,0060) DS [120]                                    #   4, 1 KVP
(0018,0088) DS [0.7]                                    #   4, 1 SpacingBetweenS
lices
(0018,0090) DS [250]                                    #   4, 1 DataCollectionD
iameter
(0018,1020) LO [2.5]                                    #   4, 1 SoftwareVersion
s
(0018,1030) LO [ANGIO CT BPU 4/Head/Hx]                 #  22, 1 ProtocolName
(0018,1040) LO [IV]                                     #   2, 1 ContrastBolusRo
ute
(0018,1100) DS [250]                                    #   4, 1 ReconstructionD
iameter
(0018,1120) DS [0]                                      #   2, 1 GantryDetectorT
ilt
(0018,1130) DS [172]                                    #   4, 1 TableHeight
(0018,1140) CS [CW]                                     #   2, 1 RotationDirecti
on
(0018,1151) IS [221]                                    #   4, 1 XRayTubeCurrent

(0018,1152) IS [250]                                    #   4, 1 Exposure
(0018,1160) SH [UB]                                     #   2, 1 FilterType
(0018,1210) SH [UB]                                     #   2, 1 ConvolutionKern
el
(0018,5100) CS [HFS]                                    #   4, 1 PatientPosition

(0020,000d) UI [1.2.840.113704.1.111.5600.1107858801.1] #  38, 1 StudyInstanceUI
D
(0020,000e) UI [1.2.840.113704.1.111.5600.1107859570.17] #  40, 1 SeriesInstance
UID
(0020,0010) SH [1963943]                                #   8, 1 StudyID
(0020,0011) IS [9712]                                   #   4, 1 SeriesNumber
(0020,0013) IS [1]                                      #   2, 1 InstanceNumber
(0020,0032) DS [-125\-63\-858]                          #  14, 3 ImagePositionPa
tient
(0020,0037) DS [1\0\0\0\1\0]                            #  12, 6 ImageOrientatio
nPatient
(0020,0052) UI [1.2.840.113704.1.111.5600.1107859163.4] #  38, 1 FrameOfReferenc
eUID
(0020,1041) DS [-379.50]                                #   8, 1 SliceLocation
(0020,4000) LT [ANGIO CT]                               #   8, 1 ImageComments
(0028,0002) US 1                                        #   2, 1 SamplesPerPixel

(0028,0004) CS [MONOCHROME2]                            #  12, 1 PhotometricInte
rpretation
(0028,0010) US 512                                      #   2, 1 Rows
(0028,0011) US 512                                      #   2, 1 Columns
(0028,0030) DS [0.48828125\0.48828125]                  #  22, 2 PixelSpacing
(0028,0100) US 16                                       #   2, 1 BitsAllocated
(0028,0101) US 12                                       #   2, 1 BitsStored
(0028,0102) US 11                                       #   2, 1 HighBit
(0028,0103) US 0                                        #   2, 1 PixelRepresenta
tion
(0028,1050) DS [00100\00100]                            #  12, 2 WindowCenter
(0028,1051) DS [00500\00500]                            #  12, 2 WindowWidth
(0028,1052) DS [-1000]                                  #   6, 1 RescaleIntercep
t
(0028,1053) DS [1]                                      #   2, 1 RescaleSlope
(0028,2110) CS [01]                                     #   2, 1 LossyImageCompr
ession
(0028,2112) DS [5.897105]                               #   8, 1 LossyImageCompr
essionRatio
(0032,1032) PN [NULL DUMCOB]                            #  12, 1 RequestingPhysi
cian
(0032,1060) LO [CER/CT]                                 #   6, 1 RequestedProced
ureDescription
(0038,0010) LO [00.9650912]                             #  10, 1 AdmissionID
(0040,0007) LO [CER/CT]                                 #   6, 1 ScheduledProced
ureStepDescription
(0040,0254) LO [CER/CT]                                 #   6, 1 PerformedProced
ureStepDescription
(0040,1001) SH [1963943]                                #   8, 1 RequestedProced
ureID
(7fe0,0010) OB (PixelSequence #=2)                      # u/l, 1 PixelData
  (fffe,e000) pi 00\00\00\00                              #   4, 1 Item
  (fffe,e000) pi ff\4f\ff\51\00\29\00\00\00\00\02\00\00\00\02\00\00\00\00\00\00\
00... # 88906, 1 Item
(fffe,e0dd) na (SequenceDelimitationItem)               #   0, 0 SequenceDelimit
ationItem


storescu with -d (only part because cmd is too short

Code: Select all

D:     Proposed SCP/SCU Role: Default
D:     Accepted SCP/SCU Role: Default
D:   Context ID:        131 (Accepted)
D:     Abstract Syntax: =MultiframeSingleBitSecondaryCaptureImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Accepted SCP/SCU Role: Default
D:     Accepted Transfer Syntax: =LittleEndianImplicit
D:   Context ID:        133 (Transfer Syntaxes Not Supported)
D:     Abstract Syntax: =MultiframeTrueColorSecondaryCaptureImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Accepted SCP/SCU Role: Default
D:   Context ID:        135 (Accepted)
D:     Abstract Syntax: =MultiframeTrueColorSecondaryCaptureImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Accepted SCP/SCU Role: Default
D:     Accepted Transfer Syntax: =LittleEndianImplicit
D:   Context ID:        137 (Transfer Syntaxes Not Supported)
D:     Abstract Syntax: =NuclearMedicineImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Accepted SCP/SCU Role: Default
D:   Context ID:        139 (Accepted)
D:     Abstract Syntax: =NuclearMedicineImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Accepted SCP/SCU Role: Default
D:     Accepted Transfer Syntax: =LittleEndianImplicit
D:   Context ID:        141 (Abstract Syntax Not Supported)
D:     Abstract Syntax: =OphthalmicPhotography16BitImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Accepted SCP/SCU Role: Default
D:   Context ID:        143 (Abstract Syntax Not Supported)
D:     Abstract Syntax: =OphthalmicPhotography16BitImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Accepted SCP/SCU Role: Default
D:   Context ID:        145 (Abstract Syntax Not Supported)
D:     Abstract Syntax: =OphthalmicPhotography8BitImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Accepted SCP/SCU Role: Default
D:   Context ID:        147 (Abstract Syntax Not Supported)
D:     Abstract Syntax: =OphthalmicPhotography8BitImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Accepted SCP/SCU Role: Default
D:   Context ID:        149 (Abstract Syntax Not Supported)
D:     Abstract Syntax: =OphthalmicTomographyImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Accepted SCP/SCU Role: Default
D:   Context ID:        151 (Abstract Syntax Not Supported)
D:     Abstract Syntax: =OphthalmicTomographyImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Accepted SCP/SCU Role: Default
D:   Context ID:        153 (Transfer Syntaxes Not Supported)
D:     Abstract Syntax: =PositronEmissionTomographyImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Accepted SCP/SCU Role: Default
D:   Context ID:        155 (Accepted)
D:     Abstract Syntax: =PositronEmissionTomographyImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Accepted SCP/SCU Role: Default
D:     Accepted Transfer Syntax: =LittleEndianImplicit
D:   Context ID:        157 (Abstract Syntax Not Supported)
D:     Abstract Syntax: =ProcedureLogStorage
D:     Proposed SCP/SCU Role: Default
D:     Accepted SCP/SCU Role: Default
D:   Context ID:        159 (Abstract Syntax Not Supported)
D:     Abstract Syntax: =ProcedureLogStorage
D:     Proposed SCP/SCU Role: Default
D:     Accepted SCP/SCU Role: Default
D:   Context ID:        161 (Abstract Syntax Not Supported)
D:     Abstract Syntax: =PseudoColorSoftcopyPresentationStateStorage
D:     Proposed SCP/SCU Role: Default
D:     Accepted SCP/SCU Role: Default
D:   Context ID:        163 (Abstract Syntax Not Supported)
D:     Abstract Syntax: =PseudoColorSoftcopyPresentationStateStorage
D:     Proposed SCP/SCU Role: Default
D:     Accepted SCP/SCU Role: Default
D:   Context ID:        165 (Abstract Syntax Not Supported)
D:     Abstract Syntax: =RawDataStorage
D:     Proposed SCP/SCU Role: Default
D:     Accepted SCP/SCU Role: Default
D:   Context ID:        167 (Abstract Syntax Not Supported)
D:     Abstract Syntax: =RawDataStorage
D:     Proposed SCP/SCU Role: Default
D:     Accepted SCP/SCU Role: Default
D:   Context ID:        169 (Abstract Syntax Not Supported)
D:     Abstract Syntax: =RealWorldValueMappingStorage
D:     Proposed SCP/SCU Role: Default
D:     Accepted SCP/SCU Role: Default
D:   Context ID:        171 (Abstract Syntax Not Supported)
D:     Abstract Syntax: =RealWorldValueMappingStorage
D:     Proposed SCP/SCU Role: Default
D:     Accepted SCP/SCU Role: Default
D:   Context ID:        173 (Transfer Syntaxes Not Supported)
D:     Abstract Syntax: =RTBeamsTreatmentRecordStorage
D:     Proposed SCP/SCU Role: Default
D:     Accepted SCP/SCU Role: Default
D:   Context ID:        175 (Accepted)
D:     Abstract Syntax: =RTBeamsTreatmentRecordStorage
D:     Proposed SCP/SCU Role: Default
D:     Accepted SCP/SCU Role: Default
D:     Accepted Transfer Syntax: =LittleEndianImplicit
D:   Context ID:        177 (Transfer Syntaxes Not Supported)
D:     Abstract Syntax: =RTBrachyTreatmentRecordStorage
D:     Proposed SCP/SCU Role: Default
D:     Accepted SCP/SCU Role: Default
D:   Context ID:        179 (Accepted)
D:     Abstract Syntax: =RTBrachyTreatmentRecordStorage
D:     Proposed SCP/SCU Role: Default
D:     Accepted SCP/SCU Role: Default
D:     Accepted Transfer Syntax: =LittleEndianImplicit
D:   Context ID:        181 (Transfer Syntaxes Not Supported)
D:     Abstract Syntax: =RTDoseStorage
D:     Proposed SCP/SCU Role: Default
D:     Accepted SCP/SCU Role: Default
D:   Context ID:        183 (Accepted)
D:     Abstract Syntax: =RTDoseStorage
D:     Proposed SCP/SCU Role: Default
D:     Accepted SCP/SCU Role: Default
D:     Accepted Transfer Syntax: =LittleEndianImplicit
D:   Context ID:        185 (Transfer Syntaxes Not Supported)
D:     Abstract Syntax: =RTImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Accepted SCP/SCU Role: Default
D:   Context ID:        187 (Accepted)
D:     Abstract Syntax: =RTImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Accepted SCP/SCU Role: Default
D:     Accepted Transfer Syntax: =LittleEndianImplicit
D:   Context ID:        189 (Transfer Syntaxes Not Supported)
D:     Abstract Syntax: =RTPlanStorage
D:     Proposed SCP/SCU Role: Default
D:     Accepted SCP/SCU Role: Default
D:   Context ID:        191 (Accepted)
D:     Abstract Syntax: =RTPlanStorage
D:     Proposed SCP/SCU Role: Default
D:     Accepted SCP/SCU Role: Default
D:     Accepted Transfer Syntax: =LittleEndianImplicit
D:   Context ID:        193 (Transfer Syntaxes Not Supported)
D:     Abstract Syntax: =RTStructureSetStorage
D:     Proposed SCP/SCU Role: Default
D:     Accepted SCP/SCU Role: Default
D:   Context ID:        195 (Accepted)
D:     Abstract Syntax: =RTStructureSetStorage
D:     Proposed SCP/SCU Role: Default
D:     Accepted SCP/SCU Role: Default
D:     Accepted Transfer Syntax: =LittleEndianImplicit
D:   Context ID:        197 (Transfer Syntaxes Not Supported)
D:     Abstract Syntax: =RTTreatmentSummaryRecordStorage
D:     Proposed SCP/SCU Role: Default
D:     Accepted SCP/SCU Role: Default
D:   Context ID:        199 (Accepted)
D:     Abstract Syntax: =RTTreatmentSummaryRecordStorage
D:     Proposed SCP/SCU Role: Default
D:     Accepted SCP/SCU Role: Default
D:     Accepted Transfer Syntax: =LittleEndianImplicit
D:   Context ID:        201 (Transfer Syntaxes Not Supported)
D:     Abstract Syntax: =SecondaryCaptureImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Accepted SCP/SCU Role: Default
D:   Context ID:        203 (Accepted)
D:     Abstract Syntax: =SecondaryCaptureImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Accepted SCP/SCU Role: Default
D:     Accepted Transfer Syntax: =LittleEndianImplicit
D:   Context ID:        205 (Abstract Syntax Not Supported)
D:     Abstract Syntax: =SpatialFiducialsStorage
D:     Proposed SCP/SCU Role: Default
D:     Accepted SCP/SCU Role: Default
D:   Context ID:        207 (Abstract Syntax Not Supported)
D:     Abstract Syntax: =SpatialFiducialsStorage
D:     Proposed SCP/SCU Role: Default
D:     Accepted SCP/SCU Role: Default
D:   Context ID:        209 (Abstract Syntax Not Supported)
D:     Abstract Syntax: =SpatialRegistrationStorage
D:     Proposed SCP/SCU Role: Default
D:     Accepted SCP/SCU Role: Default
D:   Context ID:        211 (Abstract Syntax Not Supported)
D:     Abstract Syntax: =SpatialRegistrationStorage
D:     Proposed SCP/SCU Role: Default
D:     Accepted SCP/SCU Role: Default
D:   Context ID:        213 (Abstract Syntax Not Supported)
D:     Abstract Syntax: =StereometricRelationshipStorage
D:     Proposed SCP/SCU Role: Default
D:     Accepted SCP/SCU Role: Default
D:   Context ID:        215 (Abstract Syntax Not Supported)
D:     Abstract Syntax: =StereometricRelationshipStorage
D:     Proposed SCP/SCU Role: Default
D:     Accepted SCP/SCU Role: Default
D:   Context ID:        217 (Abstract Syntax Not Supported)
D:     Abstract Syntax: =TwelveLeadECGWaveformStorage
D:     Proposed SCP/SCU Role: Default
D:     Accepted SCP/SCU Role: Default
D:   Context ID:        219 (Abstract Syntax Not Supported)
D:     Abstract Syntax: =TwelveLeadECGWaveformStorage
D:     Proposed SCP/SCU Role: Default
D:     Accepted SCP/SCU Role: Default
D:   Context ID:        221 (Transfer Syntaxes Not Supported)
D:     Abstract Syntax: =UltrasoundImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Accepted SCP/SCU Role: Default
D:   Context ID:        223 (Accepted)
D:     Abstract Syntax: =UltrasoundImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Accepted SCP/SCU Role: Default
D:     Accepted Transfer Syntax: =LittleEndianImplicit
D:   Context ID:        225 (Transfer Syntaxes Not Supported)
D:     Abstract Syntax: =UltrasoundMultiframeImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Accepted SCP/SCU Role: Default
D:   Context ID:        227 (Accepted)
D:     Abstract Syntax: =UltrasoundMultiframeImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Accepted SCP/SCU Role: Default
D:     Accepted Transfer Syntax: =LittleEndianImplicit
D:   Context ID:        229 (Transfer Syntaxes Not Supported)
D:     Abstract Syntax: =VLEndoscopicImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Accepted SCP/SCU Role: Default
D:   Context ID:        231 (Accepted)
D:     Abstract Syntax: =VLEndoscopicImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Accepted SCP/SCU Role: Default
D:     Accepted Transfer Syntax: =LittleEndianImplicit
D:   Context ID:        233 (Transfer Syntaxes Not Supported)
D:     Abstract Syntax: =VLMicroscopicImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Accepted SCP/SCU Role: Default
D:   Context ID:        235 (Accepted)
D:     Abstract Syntax: =VLMicroscopicImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Accepted SCP/SCU Role: Default
D:     Accepted Transfer Syntax: =LittleEndianImplicit
D:   Context ID:        237 (Transfer Syntaxes Not Supported)
D:     Abstract Syntax: =VLPhotographicImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Accepted SCP/SCU Role: Default
D:   Context ID:        239 (Accepted)
D:     Abstract Syntax: =VLPhotographicImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Accepted SCP/SCU Role: Default
D:     Accepted Transfer Syntax: =LittleEndianImplicit
D:   Context ID:        241 (Transfer Syntaxes Not Supported)
D:     Abstract Syntax: =VLSlideCoordinatesMicroscopicImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Accepted SCP/SCU Role: Default
D:   Context ID:        243 (Accepted)
D:     Abstract Syntax: =VLSlideCoordinatesMicroscopicImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Accepted SCP/SCU Role: Default
D:     Accepted Transfer Syntax: =LittleEndianImplicit
D:   Context ID:        245 (Transfer Syntaxes Not Supported)
D:     Abstract Syntax: =XRayAngiographicImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Accepted SCP/SCU Role: Default
D:   Context ID:        247 (Accepted)
D:     Abstract Syntax: =XRayAngiographicImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Accepted SCP/SCU Role: Default
D:     Accepted Transfer Syntax: =LittleEndianImplicit
D:   Context ID:        249 (Abstract Syntax Not Supported)
D:     Abstract Syntax: =XRayRadiationDoseSRStorage
D:     Proposed SCP/SCU Role: Default
D:     Accepted SCP/SCU Role: Default
D:   Context ID:        251 (Abstract Syntax Not Supported)
D:     Abstract Syntax: =XRayRadiationDoseSRStorage
D:     Proposed SCP/SCU Role: Default
D:     Accepted SCP/SCU Role: Default
D:   Context ID:        253 (Transfer Syntaxes Not Supported)
D:     Abstract Syntax: =XRayRadiofluoroscopicImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Accepted SCP/SCU Role: Default
D:   Context ID:        255 (Accepted)
D:     Abstract Syntax: =XRayRadiofluoroscopicImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Accepted SCP/SCU Role: Default
D:     Accepted Transfer Syntax: =LittleEndianImplicit
D: Requested Extended Negotiation: none
D: Accepted Extended Negotiation:  none
D: Requested User Identity Negotiation: none
D: User Identity Negotiation Response:  none
D: ======================= END A-ASSOCIATE-AC ======================
I: Association Accepted (Max Send PDV: 16372)
I: Sending file: c:\ll.dcm
D: DcmItem::checkTransferSyntax() TransferSyntax="Little Endian Explicit"
I: Transfer Syntax: JPEG2000 -> LittleEndianImplicit
I: Sending Store Request: MsgID 1, (CT)
D: ===================== OUTGOING DIMSE MESSAGE ====================
D: Message Type                  : C-STORE RQ
D: Message ID                    : 1
D: Affected SOP Class UID        : CTImageStorage
D: Affected SOP Instance UID     : 1.2.840.113704.1.111.3352.1107859679.1269
D: Data Set                      : present
D: Priority                      : low
D: ======================= END DIMSE MESSAGE =======================
W: DIMSE Warning: (STORESCU,ANY-SCP): sendMessage: unable to convert dataset fro
m 'JPEG 2000 (Lossless or Lossy)' transfer syntax to 'Little Endian Implicit'
E: Store Failed, file: c:\ll.dcm:
E: 0006:020e DIMSE Failed to send message
E: Store SCU Failed: 0006:020e DIMSE Failed to send message
I: Aborting Association

Michael Onken
DCMTK Developer
Posts: 1815
Joined: Fri, 2004-11-05, 13:47
Location: Oldenburg, Germany
Contact:

#9 Post by Michael Onken » Fri, 2011-05-27, 09:14

Hi,

the interesting part is actually missing in the storescu log ;)
Please quote all blocks that refer to CT Image Storage.

Also, as I said, use options -xv and -xw. This will propose JPEG200 to the server.

If your PACS does not support JPEG2000 at all, you will not be able to send it with storescu since storescu cannot decompress the file.
In that case you have to decompress it manually using a tool from another toolkit (or buy/evaluate DCMJP2K).

Michael

Forward
Posts: 17
Joined: Mon, 2011-05-02, 11:54

#10 Post by Forward » Fri, 2011-05-27, 09:19

The problem is that the command line clear the text from above, i will try it again and post the code ;)

Code: Select all

D: ====================== BEGIN A-ASSOCIATE-RQ =====================
D: Our Implementation Class UID:      1.2.276.0.7230010.3.0.3.6.0
D: Our Implementation Version Name:   OFFIS_DCMTK_360
D: Their Implementation Class UID:    
D: Their Implementation Version Name: 
D: Application Context Name:    1.2.840.10008.3.1.1.1
D: Calling Application Name:    STORESCU
D: Called Application Name:     ANY-SCP
D: Responding Application Name: resp. AP Title
D: Our Max PDU Receive Size:    16384
D: Their Max PDU Receive Size:  0
D: Presentation Contexts:
D:   Context ID:        1 (Proposed)
D:     Abstract Syntax: =AmbulatoryECGWaveformStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        3 (Proposed)
D:     Abstract Syntax: =AmbulatoryECGWaveformStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        5 (Proposed)
D:     Abstract Syntax: =BasicTextSRStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        7 (Proposed)
D:     Abstract Syntax: =BasicTextSRStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        9 (Proposed)
D:     Abstract Syntax: =BasicVoiceAudioWaveformStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        11 (Proposed)
D:     Abstract Syntax: =BasicVoiceAudioWaveformStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        13 (Proposed)
D:     Abstract Syntax: =BlendingSoftcopyPresentationStateStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        15 (Proposed)
D:     Abstract Syntax: =BlendingSoftcopyPresentationStateStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        17 (Proposed)
D:     Abstract Syntax: =CardiacElectrophysiologyWaveformStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        19 (Proposed)
D:     Abstract Syntax: =CardiacElectrophysiologyWaveformStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        21 (Proposed)
D:     Abstract Syntax: =ChestCADSRStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        23 (Proposed)
D:     Abstract Syntax: =ChestCADSRStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        25 (Proposed)
D:     Abstract Syntax: =ColonCADSRStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        27 (Proposed)
D:     Abstract Syntax: =ColonCADSRStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        29 (Proposed)
D:     Abstract Syntax: =ColorSoftcopyPresentationStateStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        31 (Proposed)
D:     Abstract Syntax: =ColorSoftcopyPresentationStateStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        33 (Proposed)
D:     Abstract Syntax: =ComprehensiveSRStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        35 (Proposed)
D:     Abstract Syntax: =ComprehensiveSRStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        37 (Proposed)
D:     Abstract Syntax: =ComputedRadiographyImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        39 (Proposed)
D:     Abstract Syntax: =ComputedRadiographyImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        41 (Proposed)
D:     Abstract Syntax: =CTImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        43 (Proposed)
D:     Abstract Syntax: =CTImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        45 (Proposed)
D:     Abstract Syntax: =DigitalIntraOralXRayImageStorageForPresentation
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        47 (Proposed)
D:     Abstract Syntax: =DigitalIntraOralXRayImageStorageForPresentation
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        49 (Proposed)
D:     Abstract Syntax: =DigitalIntraOralXRayImageStorageForProcessing
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        51 (Proposed)
D:     Abstract Syntax: =DigitalIntraOralXRayImageStorageForProcessing
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        53 (Proposed)
D:     Abstract Syntax: =DigitalMammographyXRayImageStorageForPresentation
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        55 (Proposed)
D:     Abstract Syntax: =DigitalMammographyXRayImageStorageForPresentation
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        57 (Proposed)
D:     Abstract Syntax: =DigitalMammographyXRayImageStorageForProcessing
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        59 (Proposed)
D:     Abstract Syntax: =DigitalMammographyXRayImageStorageForProcessing
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        61 (Proposed)
D:     Abstract Syntax: =DigitalXRayImageStorageForPresentation
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        63 (Proposed)
D:     Abstract Syntax: =DigitalXRayImageStorageForPresentation
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        65 (Proposed)
D:     Abstract Syntax: =DigitalXRayImageStorageForProcessing
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        67 (Proposed)
D:     Abstract Syntax: =DigitalXRayImageStorageForProcessing
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        69 (Proposed)
D:     Abstract Syntax: =EncapsulatedPDFStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        71 (Proposed)
D:     Abstract Syntax: =EncapsulatedPDFStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        73 (Proposed)
D:     Abstract Syntax: =EnhancedCTImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        75 (Proposed)
D:     Abstract Syntax: =EnhancedCTImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        77 (Proposed)
D:     Abstract Syntax: =EnhancedMRImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        79 (Proposed)
D:     Abstract Syntax: =EnhancedMRImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        81 (Proposed)
D:     Abstract Syntax: =EnhancedSRStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        83 (Proposed)
D:     Abstract Syntax: =EnhancedSRStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        85 (Proposed)
D:     Abstract Syntax: =EnhancedXAImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        87 (Proposed)
D:     Abstract Syntax: =EnhancedXAImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        89 (Proposed)
D:     Abstract Syntax: =EnhancedXRFImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        91 (Proposed)
D:     Abstract Syntax: =EnhancedXRFImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        93 (Proposed)
D:     Abstract Syntax: =GeneralECGWaveformStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        95 (Proposed)
D:     Abstract Syntax: =GeneralECGWaveformStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        97 (Proposed)
D:     Abstract Syntax: =GrayscaleSoftcopyPresentationStateStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        99 (Proposed)
D:     Abstract Syntax: =GrayscaleSoftcopyPresentationStateStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        101 (Proposed)
D:     Abstract Syntax: =HemodynamicWaveformStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        103 (Proposed)
D:     Abstract Syntax: =HemodynamicWaveformStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        105 (Proposed)
D:     Abstract Syntax: =KeyObjectSelectionDocumentStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        107 (Proposed)
D:     Abstract Syntax: =KeyObjectSelectionDocumentStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        109 (Proposed)
D:     Abstract Syntax: =MammographyCADSRStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        111 (Proposed)
D:     Abstract Syntax: =MammographyCADSRStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        113 (Proposed)
D:     Abstract Syntax: =MRImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        115 (Proposed)
D:     Abstract Syntax: =MRImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        117 (Proposed)
D:     Abstract Syntax: =MRSpectroscopyStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        119 (Proposed)
D:     Abstract Syntax: =MRSpectroscopyStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        121 (Proposed)
D:     Abstract Syntax: =MultiframeGrayscaleByteSecondaryCaptureImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        123 (Proposed)
D:     Abstract Syntax: =MultiframeGrayscaleByteSecondaryCaptureImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        125 (Proposed)
D:     Abstract Syntax: =MultiframeGrayscaleWordSecondaryCaptureImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        127 (Proposed)
D:     Abstract Syntax: =MultiframeGrayscaleWordSecondaryCaptureImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        129 (Proposed)
D:     Abstract Syntax: =MultiframeSingleBitSecondaryCaptureImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        131 (Proposed)
D:     Abstract Syntax: =MultiframeSingleBitSecondaryCaptureImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        133 (Proposed)
D:     Abstract Syntax: =MultiframeTrueColorSecondaryCaptureImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        135 (Proposed)
D:     Abstract Syntax: =MultiframeTrueColorSecondaryCaptureImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        137 (Proposed)
D:     Abstract Syntax: =NuclearMedicineImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        139 (Proposed)
D:     Abstract Syntax: =NuclearMedicineImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        141 (Proposed)
D:     Abstract Syntax: =OphthalmicPhotography16BitImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        143 (Proposed)
D:     Abstract Syntax: =OphthalmicPhotography16BitImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        145 (Proposed)
D:     Abstract Syntax: =OphthalmicPhotography8BitImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        147 (Proposed)
D:     Abstract Syntax: =OphthalmicPhotography8BitImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        149 (Proposed)
D:     Abstract Syntax: =OphthalmicTomographyImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        151 (Proposed)
D:     Abstract Syntax: =OphthalmicTomographyImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        153 (Proposed)
D:     Abstract Syntax: =PositronEmissionTomographyImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        155 (Proposed)
D:     Abstract Syntax: =PositronEmissionTomographyImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        157 (Proposed)
D:     Abstract Syntax: =ProcedureLogStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        159 (Proposed)
D:     Abstract Syntax: =ProcedureLogStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        161 (Proposed)
D:     Abstract Syntax: =PseudoColorSoftcopyPresentationStateStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        163 (Proposed)
D:     Abstract Syntax: =PseudoColorSoftcopyPresentationStateStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        165 (Proposed)
D:     Abstract Syntax: =RawDataStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        167 (Proposed)
D:     Abstract Syntax: =RawDataStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        169 (Proposed)
D:     Abstract Syntax: =RealWorldValueMappingStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        171 (Proposed)
D:     Abstract Syntax: =RealWorldValueMappingStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        173 (Proposed)
D:     Abstract Syntax: =RTBeamsTreatmentRecordStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        175 (Proposed)
D:     Abstract Syntax: =RTBeamsTreatmentRecordStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        177 (Proposed)
D:     Abstract Syntax: =RTBrachyTreatmentRecordStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        179 (Proposed)
D:     Abstract Syntax: =RTBrachyTreatmentRecordStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        181 (Proposed)
D:     Abstract Syntax: =RTDoseStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        183 (Proposed)
D:     Abstract Syntax: =RTDoseStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        185 (Proposed)
D:     Abstract Syntax: =RTImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        187 (Proposed)
D:     Abstract Syntax: =RTImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        189 (Proposed)
D:     Abstract Syntax: =RTPlanStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        191 (Proposed)
D:     Abstract Syntax: =RTPlanStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        193 (Proposed)
D:     Abstract Syntax: =RTStructureSetStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        195 (Proposed)
D:     Abstract Syntax: =RTStructureSetStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        197 (Proposed)
D:     Abstract Syntax: =RTTreatmentSummaryRecordStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        199 (Proposed)
D:     Abstract Syntax: =RTTreatmentSummaryRecordStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        201 (Proposed)
D:     Abstract Syntax: =SecondaryCaptureImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        203 (Proposed)
D:     Abstract Syntax: =SecondaryCaptureImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        205 (Proposed)
D:     Abstract Syntax: =SpatialFiducialsStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        207 (Proposed)
D:     Abstract Syntax: =SpatialFiducialsStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        209 (Proposed)
D:     Abstract Syntax: =SpatialRegistrationStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        211 (Proposed)
D:     Abstract Syntax: =SpatialRegistrationStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        213 (Proposed)
D:     Abstract Syntax: =StereometricRelationshipStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        215 (Proposed)
D:     Abstract Syntax: =StereometricRelationshipStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        217 (Proposed)
D:     Abstract Syntax: =TwelveLeadECGWaveformStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        219 (Proposed)
D:     Abstract Syntax: =TwelveLeadECGWaveformStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        221 (Proposed)
D:     Abstract Syntax: =UltrasoundImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        223 (Proposed)
D:     Abstract Syntax: =UltrasoundImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        225 (Proposed)
D:     Abstract Syntax: =UltrasoundMultiframeImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        227 (Proposed)
D:     Abstract Syntax: =UltrasoundMultiframeImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        229 (Proposed)
D:     Abstract Syntax: =VLEndoscopicImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        231 (Proposed)
D:     Abstract Syntax: =VLEndoscopicImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        233 (Proposed)
D:     Abstract Syntax: =VLMicroscopicImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        235 (Proposed)
D:     Abstract Syntax: =VLMicroscopicImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        237 (Proposed)
D:     Abstract Syntax: =VLPhotographicImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        239 (Proposed)
D:     Abstract Syntax: =VLPhotographicImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        241 (Proposed)
D:     Abstract Syntax: =VLSlideCoordinatesMicroscopicImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        243 (Proposed)
D:     Abstract Syntax: =VLSlideCoordinatesMicroscopicImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        245 (Proposed)
D:     Abstract Syntax: =XRayAngiographicImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        247 (Proposed)
D:     Abstract Syntax: =XRayAngiographicImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        249 (Proposed)
D:     Abstract Syntax: =XRayRadiationDoseSRStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        251 (Proposed)
D:     Abstract Syntax: =XRayRadiationDoseSRStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        253 (Proposed)
D:     Abstract Syntax: =XRayRadiofluoroscopicImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        255 (Proposed)
D:     Abstract Syntax: =XRayRadiofluoroscopicImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D: Requested Extended Negotiation: none
D: Accepted Extended Negotiation:  none
D: Requested User Identity Negotiation: none
D: User Identity Negotiation Response:  none
D: ======================= END A-ASSOCIATE-RQ ======================
I: Requesting Association
D: Constructing Associate RQ PDU
D: PDU Type: Associate Accept, PDU Length: 3854 + 6 bytes PDU header
D: Only dumping 512 bytes.
D:   02  00  00  00  0f  0e  00  01  00  00  41  4e  59  2d  53  43
D:   50  20  20  20  20  20  20  20  20  20  53  54  4f  52  45  53
D:   43  55  20  20  20  20  20  20  20  20  00  00  00  00  00  00
D:   00  00  00  00  00  00  00  00  00  00  00  00  00  00  00  00
D:   00  00  00  00  00  00  00  00  00  00  10  00  00  15  31  2e
D:   32  2e  38  34  30  2e  31  30  30  30  38  2e  33  2e  31  2e
D:   31  2e  31  21  00  00  08  01  00  03  00  40  00  00  00  21
D:   00  00  19  03  00  03  00  40  00  00  11  31  2e  32  2e  38
D:   34  30  2e  31  30  30  30  38  2e  31  2e  32  21  00  00  19
D:   05  00  04  00  40  00  00  11  31  2e  32  2e  38  34  30  2e
D:   31  30  30  30  38  2e  31  2e  32  21  00  00  19  07  00  00
D:   00  40  00  00  11  31  2e  32  2e  38  34  30  2e  31  30  30
D:   30  38  2e  31  2e  32  21  00  00  19  09  00  03  00  40  00
D:   00  11  31  2e  32  2e  38  34  30  2e  31  30  30  30  38  2e
D:   31  2e  32  21  00  00  19  0b  00  03  00  40  00  00  11  31
D:   2e  32  2e  38  34  30  2e  31  30  30  30  38  2e  31  2e  32
D:   21  00  00  19  0d  00  03  00  40  00  00  11  31  2e  32  2e
D:   38  34  30  2e  31  30  30  30  38  2e  31  2e  32  21  00  00
D:   19  0f  00  03  00  40  00  00  11  31  2e  32  2e  38  34  30
D:   2e  31  30  30  30  38  2e  31  2e  32  21  00  00  19  11  00
D:   03  00  40  00  00  11  31  2e  32  2e  38  34  30  2e  31  30
D:   30  30  38  2e  31  2e  32  21  00  00  19  13  00  03  00  40
D:   00  00  11  31  2e  32  2e  38  34  30  2e  31  30  30  30  38
D:   2e  31  2e  32  21  00  00  19  15  00  03  00  40  00  00  11
D:   31  2e  32  2e  38  34  30  2e  31  30  30  30  38  2e  31  2e
D:   32  21  00  00  19  17  00  03  00  40  00  00  11  31  2e  32
D:   2e  38  34  30  2e  31  30  30  30  38  2e  31  2e  32  21  00
D:   00  19  19  00  03  00  40  00  00  11  31  2e  32  2e  38  34
D:   30  2e  31  30  30  30  38  2e  31  2e  32  21  00  00  19  1b
D:   00  03  00  40  00  00  11  31  2e  32  2e  38  34  30  2e  31
D:   30  30  30  38  2e  31  2e  32  21  00  00  19  1d  00  03  00
D:   40  00  00  11  31  2e  32  2e  38  34  30  2e  31  30  30  30
D: 
D: Parsing an A-ASSOCIATE PDU
D: Transport connection: TCP/IP, unencrypted.
D: Association Parameters Negotiated:
D: ====================== BEGIN A-ASSOCIATE-AC =====================
D: Our Implementation Class UID:      1.2.276.0.7230010.3.0.3.6.0
D: Our Implementation Version Name:   OFFIS_DCMTK_360
D: Their Implementation Class UID:    1.2.826.0.1.3680043.2.135.1066.101
D: Their Implementation Version Name: 1.4.16/WIN32
D: Application Context Name:    1.2.840.10008.3.1.1.1
D: Calling Application Name:    STORESCU
D: Called Application Name:     ANY-SCP
D: Responding Application Name: ANY-SCP
D: Our Max PDU Receive Size:    16384
D: Their Max PDU Receive Size:  16384
D: Presentation Contexts:
D:   Context ID:        1 (Abstract Syntax Not Supported)
D:     Abstract Syntax: =AmbulatoryECGWaveformStorage
D:     Proposed SCP/SCU Role: Default
D:     Accepted SCP/SCU Role: Default
D:   Context ID:        3 (Abstract Syntax Not Supported)

Forward
Posts: 17
Joined: Mon, 2011-05-02, 11:54

#11 Post by Forward » Fri, 2011-05-27, 09:46

The problem is that the command line clear the text from above, i will try it again and post the code ;)

Code: Select all

D: ====================== BEGIN A-ASSOCIATE-RQ =====================
D: Our Implementation Class UID:      1.2.276.0.7230010.3.0.3.6.0
D: Our Implementation Version Name:   OFFIS_DCMTK_360
D: Their Implementation Class UID:    
D: Their Implementation Version Name: 
D: Application Context Name:    1.2.840.10008.3.1.1.1
D: Calling Application Name:    STORESCU
D: Called Application Name:     ANY-SCP
D: Responding Application Name: resp. AP Title
D: Our Max PDU Receive Size:    16384
D: Their Max PDU Receive Size:  0
D: Presentation Contexts:
D:   Context ID:        1 (Proposed)
D:     Abstract Syntax: =AmbulatoryECGWaveformStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        3 (Proposed)
D:     Abstract Syntax: =AmbulatoryECGWaveformStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        5 (Proposed)
D:     Abstract Syntax: =BasicTextSRStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        7 (Proposed)
D:     Abstract Syntax: =BasicTextSRStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        9 (Proposed)
D:     Abstract Syntax: =BasicVoiceAudioWaveformStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        11 (Proposed)
D:     Abstract Syntax: =BasicVoiceAudioWaveformStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        13 (Proposed)
D:     Abstract Syntax: =BlendingSoftcopyPresentationStateStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        15 (Proposed)
D:     Abstract Syntax: =BlendingSoftcopyPresentationStateStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        17 (Proposed)
D:     Abstract Syntax: =CardiacElectrophysiologyWaveformStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        19 (Proposed)
D:     Abstract Syntax: =CardiacElectrophysiologyWaveformStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        21 (Proposed)
D:     Abstract Syntax: =ChestCADSRStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        23 (Proposed)
D:     Abstract Syntax: =ChestCADSRStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        25 (Proposed)
D:     Abstract Syntax: =ColonCADSRStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        27 (Proposed)
D:     Abstract Syntax: =ColonCADSRStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        29 (Proposed)
D:     Abstract Syntax: =ColorSoftcopyPresentationStateStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        31 (Proposed)
D:     Abstract Syntax: =ColorSoftcopyPresentationStateStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        33 (Proposed)
D:     Abstract Syntax: =ComprehensiveSRStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        35 (Proposed)
D:     Abstract Syntax: =ComprehensiveSRStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        37 (Proposed)
D:     Abstract Syntax: =ComputedRadiographyImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        39 (Proposed)
D:     Abstract Syntax: =ComputedRadiographyImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        41 (Proposed)
D:     Abstract Syntax: =CTImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        43 (Proposed)
D:     Abstract Syntax: =CTImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        45 (Proposed)
D:     Abstract Syntax: =DigitalIntraOralXRayImageStorageForPresentation
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        47 (Proposed)
D:     Abstract Syntax: =DigitalIntraOralXRayImageStorageForPresentation
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        49 (Proposed)
D:     Abstract Syntax: =DigitalIntraOralXRayImageStorageForProcessing
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        51 (Proposed)
D:     Abstract Syntax: =DigitalIntraOralXRayImageStorageForProcessing
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        53 (Proposed)
D:     Abstract Syntax: =DigitalMammographyXRayImageStorageForPresentation
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        55 (Proposed)
D:     Abstract Syntax: =DigitalMammographyXRayImageStorageForPresentation
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        57 (Proposed)
D:     Abstract Syntax: =DigitalMammographyXRayImageStorageForProcessing
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        59 (Proposed)
D:     Abstract Syntax: =DigitalMammographyXRayImageStorageForProcessing
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        61 (Proposed)
D:     Abstract Syntax: =DigitalXRayImageStorageForPresentation
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        63 (Proposed)
D:     Abstract Syntax: =DigitalXRayImageStorageForPresentation
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        65 (Proposed)
D:     Abstract Syntax: =DigitalXRayImageStorageForProcessing
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        67 (Proposed)
D:     Abstract Syntax: =DigitalXRayImageStorageForProcessing
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        69 (Proposed)
D:     Abstract Syntax: =EncapsulatedPDFStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        71 (Proposed)
D:     Abstract Syntax: =EncapsulatedPDFStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        73 (Proposed)
D:     Abstract Syntax: =EnhancedCTImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        75 (Proposed)
D:     Abstract Syntax: =EnhancedCTImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        77 (Proposed)
D:     Abstract Syntax: =EnhancedMRImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        79 (Proposed)
D:     Abstract Syntax: =EnhancedMRImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        81 (Proposed)
D:     Abstract Syntax: =EnhancedSRStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        83 (Proposed)
D:     Abstract Syntax: =EnhancedSRStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        85 (Proposed)
D:     Abstract Syntax: =EnhancedXAImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        87 (Proposed)
D:     Abstract Syntax: =EnhancedXAImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        89 (Proposed)
D:     Abstract Syntax: =EnhancedXRFImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        91 (Proposed)
D:     Abstract Syntax: =EnhancedXRFImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        93 (Proposed)
D:     Abstract Syntax: =GeneralECGWaveformStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        95 (Proposed)
D:     Abstract Syntax: =GeneralECGWaveformStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        97 (Proposed)
D:     Abstract Syntax: =GrayscaleSoftcopyPresentationStateStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        99 (Proposed)
D:     Abstract Syntax: =GrayscaleSoftcopyPresentationStateStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        101 (Proposed)
D:     Abstract Syntax: =HemodynamicWaveformStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        103 (Proposed)
D:     Abstract Syntax: =HemodynamicWaveformStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        105 (Proposed)
D:     Abstract Syntax: =KeyObjectSelectionDocumentStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        107 (Proposed)
D:     Abstract Syntax: =KeyObjectSelectionDocumentStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        109 (Proposed)
D:     Abstract Syntax: =MammographyCADSRStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        111 (Proposed)
D:     Abstract Syntax: =MammographyCADSRStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        113 (Proposed)
D:     Abstract Syntax: =MRImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        115 (Proposed)
D:     Abstract Syntax: =MRImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        117 (Proposed)
D:     Abstract Syntax: =MRSpectroscopyStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        119 (Proposed)
D:     Abstract Syntax: =MRSpectroscopyStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        121 (Proposed)
D:     Abstract Syntax: =MultiframeGrayscaleByteSecondaryCaptureImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        123 (Proposed)
D:     Abstract Syntax: =MultiframeGrayscaleByteSecondaryCaptureImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        125 (Proposed)
D:     Abstract Syntax: =MultiframeGrayscaleWordSecondaryCaptureImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        127 (Proposed)
D:     Abstract Syntax: =MultiframeGrayscaleWordSecondaryCaptureImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        129 (Proposed)
D:     Abstract Syntax: =MultiframeSingleBitSecondaryCaptureImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        131 (Proposed)
D:     Abstract Syntax: =MultiframeSingleBitSecondaryCaptureImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        133 (Proposed)
D:     Abstract Syntax: =MultiframeTrueColorSecondaryCaptureImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        135 (Proposed)
D:     Abstract Syntax: =MultiframeTrueColorSecondaryCaptureImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        137 (Proposed)
D:     Abstract Syntax: =NuclearMedicineImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        139 (Proposed)
D:     Abstract Syntax: =NuclearMedicineImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        141 (Proposed)
D:     Abstract Syntax: =OphthalmicPhotography16BitImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        143 (Proposed)
D:     Abstract Syntax: =OphthalmicPhotography16BitImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        145 (Proposed)
D:     Abstract Syntax: =OphthalmicPhotography8BitImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        147 (Proposed)
D:     Abstract Syntax: =OphthalmicPhotography8BitImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        149 (Proposed)
D:     Abstract Syntax: =OphthalmicTomographyImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        151 (Proposed)
D:     Abstract Syntax: =OphthalmicTomographyImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        153 (Proposed)
D:     Abstract Syntax: =PositronEmissionTomographyImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        155 (Proposed)
D:     Abstract Syntax: =PositronEmissionTomographyImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        157 (Proposed)
D:     Abstract Syntax: =ProcedureLogStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        159 (Proposed)
D:     Abstract Syntax: =ProcedureLogStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        161 (Proposed)
D:     Abstract Syntax: =PseudoColorSoftcopyPresentationStateStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        163 (Proposed)
D:     Abstract Syntax: =PseudoColorSoftcopyPresentationStateStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        165 (Proposed)
D:     Abstract Syntax: =RawDataStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        167 (Proposed)
D:     Abstract Syntax: =RawDataStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        169 (Proposed)
D:     Abstract Syntax: =RealWorldValueMappingStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        171 (Proposed)
D:     Abstract Syntax: =RealWorldValueMappingStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        173 (Proposed)
D:     Abstract Syntax: =RTBeamsTreatmentRecordStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        175 (Proposed)
D:     Abstract Syntax: =RTBeamsTreatmentRecordStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        177 (Proposed)
D:     Abstract Syntax: =RTBrachyTreatmentRecordStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        179 (Proposed)
D:     Abstract Syntax: =RTBrachyTreatmentRecordStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        181 (Proposed)
D:     Abstract Syntax: =RTDoseStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        183 (Proposed)
D:     Abstract Syntax: =RTDoseStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        185 (Proposed)
D:     Abstract Syntax: =RTImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        187 (Proposed)
D:     Abstract Syntax: =RTImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        189 (Proposed)
D:     Abstract Syntax: =RTPlanStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        191 (Proposed)
D:     Abstract Syntax: =RTPlanStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        193 (Proposed)
D:     Abstract Syntax: =RTStructureSetStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        195 (Proposed)
D:     Abstract Syntax: =RTStructureSetStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        197 (Proposed)
D:     Abstract Syntax: =RTTreatmentSummaryRecordStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        199 (Proposed)
D:     Abstract Syntax: =RTTreatmentSummaryRecordStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        201 (Proposed)
D:     Abstract Syntax: =SecondaryCaptureImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        203 (Proposed)
D:     Abstract Syntax: =SecondaryCaptureImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        205 (Proposed)
D:     Abstract Syntax: =SpatialFiducialsStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        207 (Proposed)
D:     Abstract Syntax: =SpatialFiducialsStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        209 (Proposed)
D:     Abstract Syntax: =SpatialRegistrationStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        211 (Proposed)
D:     Abstract Syntax: =SpatialRegistrationStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        213 (Proposed)
D:     Abstract Syntax: =StereometricRelationshipStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        215 (Proposed)
D:     Abstract Syntax: =StereometricRelationshipStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        217 (Proposed)
D:     Abstract Syntax: =TwelveLeadECGWaveformStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        219 (Proposed)
D:     Abstract Syntax: =TwelveLeadECGWaveformStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        221 (Proposed)
D:     Abstract Syntax: =UltrasoundImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        223 (Proposed)
D:     Abstract Syntax: =UltrasoundImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        225 (Proposed)
D:     Abstract Syntax: =UltrasoundMultiframeImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        227 (Proposed)
D:     Abstract Syntax: =UltrasoundMultiframeImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        229 (Proposed)
D:     Abstract Syntax: =VLEndoscopicImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        231 (Proposed)
D:     Abstract Syntax: =VLEndoscopicImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        233 (Proposed)
D:     Abstract Syntax: =VLMicroscopicImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        235 (Proposed)
D:     Abstract Syntax: =VLMicroscopicImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        237 (Proposed)
D:     Abstract Syntax: =VLPhotographicImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        239 (Proposed)
D:     Abstract Syntax: =VLPhotographicImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        241 (Proposed)
D:     Abstract Syntax: =VLSlideCoordinatesMicroscopicImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        243 (Proposed)
D:     Abstract Syntax: =VLSlideCoordinatesMicroscopicImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        245 (Proposed)
D:     Abstract Syntax: =XRayAngiographicImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        247 (Proposed)
D:     Abstract Syntax: =XRayAngiographicImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        249 (Proposed)
D:     Abstract Syntax: =XRayRadiationDoseSRStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        251 (Proposed)
D:     Abstract Syntax: =XRayRadiationDoseSRStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D:   Context ID:        253 (Proposed)
D:     Abstract Syntax: =XRayRadiofluoroscopicImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =JPEG2000
D:   Context ID:        255 (Proposed)
D:     Abstract Syntax: =XRayRadiofluoroscopicImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Proposed Transfer Syntax(es):
D:       =LittleEndianExplicit
D:       =BigEndianExplicit
D:       =LittleEndianImplicit
D: Requested Extended Negotiation: none
D: Accepted Extended Negotiation:  none
D: Requested User Identity Negotiation: none
D: User Identity Negotiation Response:  none
D: ======================= END A-ASSOCIATE-RQ ======================
I: Requesting Association
D: Constructing Associate RQ PDU
D: PDU Type: Associate Accept, PDU Length: 3854 + 6 bytes PDU header
D: Only dumping 512 bytes.
D:   02  00  00  00  0f  0e  00  01  00  00  41  4e  59  2d  53  43
D:   50  20  20  20  20  20  20  20  20  20  53  54  4f  52  45  53
D:   43  55  20  20  20  20  20  20  20  20  00  00  00  00  00  00
D:   00  00  00  00  00  00  00  00  00  00  00  00  00  00  00  00
D:   00  00  00  00  00  00  00  00  00  00  10  00  00  15  31  2e
D:   32  2e  38  34  30  2e  31  30  30  30  38  2e  33  2e  31  2e
D:   31  2e  31  21  00  00  08  01  00  03  00  40  00  00  00  21
D:   00  00  19  03  00  03  00  40  00  00  11  31  2e  32  2e  38
D:   34  30  2e  31  30  30  30  38  2e  31  2e  32  21  00  00  19
D:   05  00  04  00  40  00  00  11  31  2e  32  2e  38  34  30  2e
D:   31  30  30  30  38  2e  31  2e  32  21  00  00  19  07  00  00
D:   00  40  00  00  11  31  2e  32  2e  38  34  30  2e  31  30  30
D:   30  38  2e  31  2e  32  21  00  00  19  09  00  03  00  40  00
D:   00  11  31  2e  32  2e  38  34  30  2e  31  30  30  30  38  2e
D:   31  2e  32  21  00  00  19  0b  00  03  00  40  00  00  11  31
D:   2e  32  2e  38  34  30  2e  31  30  30  30  38  2e  31  2e  32
D:   21  00  00  19  0d  00  03  00  40  00  00  11  31  2e  32  2e
D:   38  34  30  2e  31  30  30  30  38  2e  31  2e  32  21  00  00
D:   19  0f  00  03  00  40  00  00  11  31  2e  32  2e  38  34  30
D:   2e  31  30  30  30  38  2e  31  2e  32  21  00  00  19  11  00
D:   03  00  40  00  00  11  31  2e  32  2e  38  34  30  2e  31  30
D:   30  30  38  2e  31  2e  32  21  00  00  19  13  00  03  00  40
D:   00  00  11  31  2e  32  2e  38  34  30  2e  31  30  30  30  38
D:   2e  31  2e  32  21  00  00  19  15  00  03  00  40  00  00  11
D:   31  2e  32  2e  38  34  30  2e  31  30  30  30  38  2e  31  2e
D:   32  21  00  00  19  17  00  03  00  40  00  00  11  31  2e  32
D:   2e  38  34  30  2e  31  30  30  30  38  2e  31  2e  32  21  00
D:   00  19  19  00  03  00  40  00  00  11  31  2e  32  2e  38  34
D:   30  2e  31  30  30  30  38  2e  31  2e  32  21  00  00  19  1b
D:   00  03  00  40  00  00  11  31  2e  32  2e  38  34  30  2e  31
D:   30  30  30  38  2e  31  2e  32  21  00  00  19  1d  00  03  00
D:   40  00  00  11  31  2e  32  2e  38  34  30  2e  31  30  30  30
D: 
D: Parsing an A-ASSOCIATE PDU
D: Transport connection: TCP/IP, unencrypted.
D: Association Parameters Negotiated:
D: ====================== BEGIN A-ASSOCIATE-AC =====================
D: Our Implementation Class UID:      1.2.276.0.7230010.3.0.3.6.0
D: Our Implementation Version Name:   OFFIS_DCMTK_360
D: Their Implementation Class UID:    1.2.826.0.1.3680043.2.135.1066.101
D: Their Implementation Version Name: 1.4.16/WIN32
D: Application Context Name:    1.2.840.10008.3.1.1.1
D: Calling Application Name:    STORESCU
D: Called Application Name:     ANY-SCP
D: Responding Application Name: ANY-SCP
D: Our Max PDU Receive Size:    16384
D: Their Max PDU Receive Size:  16384
D: Presentation Contexts:
D:   Context ID:        1 (Abstract Syntax Not Supported)
D:     Abstract Syntax: =AmbulatoryECGWaveformStorage
D:     Proposed SCP/SCU Role: Default
D:     Accepted SCP/SCU Role: Default
D:   Context ID:        3 (Abstract Syntax Not Supported)

Michael Onken
DCMTK Developer
Posts: 1815
Joined: Fri, 2004-11-05, 13:47
Location: Oldenburg, Germany
Contact:

#12 Post by Michael Onken » Fri, 2011-05-27, 10:24

Thanks,

so this is the first part we need:
D: Context ID: 41 (Proposed)
D: Abstract Syntax: =CTImageStorage
D: Proposed SCP/SCU Role: Default
D: Proposed Transfer Syntax(es):
D: =JPEG2000
D: Context ID: 43 (Proposed)
D: Abstract Syntax: =CTImageStorage
D: Proposed SCP/SCU Role: Default
D: Proposed Transfer Syntax(es):
D: =LittleEndianExplicit
D: =BigEndianExplicit
D: =LittleEndianImplicit
This means you offer the server to send in JPEG 2000 (context id 41) or any uncompressed (context ID 43) transfer syntaxes.

Now look in the debug output of storescu whether context ID 41 (this is the JPEG200 one) and 43 (not that important) were accepted. There should be some lines containing the number 41,together with the words rejected or refused or accepted, and the word CTImageStorage). This text block tell you whether the server accepted your proposal to send JPEG2000 (and uncompressed).

Michael

Forward
Posts: 17
Joined: Mon, 2011-05-02, 11:54

#13 Post by Forward » Fri, 2011-05-27, 15:10

Code: Select all

D:   Context ID:        41 (Transfer Syntaxes Not Supported)
D:     Abstract Syntax: =CTImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Accepted SCP/SCU Role: Default
D:   Context ID:        43 (Accepted)
D:     Abstract Syntax: =CTImageStorage
D:     Proposed SCP/SCU Role: Default
D:     Accepted SCP/SCU Role: Default
D:     Accepted Transfer Syntax: =LittleEndianImplicit

Michael Onken
DCMTK Developer
Posts: 1815
Joined: Fri, 2004-11-05, 13:47
Location: Oldenburg, Germany
Contact:

#14 Post by Michael Onken » Fri, 2011-05-27, 15:22

So,

as you can see context ID 41 is rejected since the server does not support JPEG2000 CT images.

Thus, storescu cannot send it to the server.

You must decompress the images (as said above) and then you can send it with storescu.