Hello all,
I came across a set of DICOM ECG waveform samples and none of my viewers can display these files except the viewer "GinkgoCADx".
When I imported these DICOM samples to a PACS, the PACS server accepted the files but cannot display the images. I used dcmftest to test one of the sample which is of the type "12 lead ECG". dcmftest said it is dicom compliant (i.e. YES was the return value). When I look at this file using dcmdump, I found that there is no entry for the SOP (7fe0,0010).
My questions are:
1/ Missing the pixel data information means this file is non-dicom compliant. Correct? Why dcmftest return a "YES" result?
2/ I can understand why most dicom viewers cannot display all the "DICOM" ECG sample files from this OEM site because of the missing tags in their samples, I wonder why a particular viewer, GinkgoCADx, can display the waveform of this file while other viewers cannot?
I can post the URL where I got these sample files if I am allow to do so.
Regards,
JL
Pixel Data
Moderator: Moderator Team
Re: Pixel Data
ECG(electro cardio graph) basically is a graph.
For plotting graph, you don't required pixel information, instead you required only co-ordinate values.
Maybe "GinkgoCADx" is plotting graph by parsing the co-ordinate values available in the DICOM ECG file.
Missing pixel data information doesn't mean that file is non-DICOM. For example, Structured Report files and Presentation State(PRE) files are valid DICOM file which normally doesn't contains pixel data information.
HTH,
Alvin
For plotting graph, you don't required pixel information, instead you required only co-ordinate values.
Maybe "GinkgoCADx" is plotting graph by parsing the co-ordinate values available in the DICOM ECG file.
Missing pixel data information doesn't mean that file is non-DICOM. For example, Structured Report files and Presentation State(PRE) files are valid DICOM file which normally doesn't contains pixel data information.
HTH,
Alvin
Who is online
Users browsing this forum: No registered users and 1 guest