Hi,
I have a problem when I transmit multi-slice CT from HERMES to “STORESCP. It transmits each slice individually. But with other programs, HERMES transmits as a multi-volume DICOM image. Could please help on this?
Regards
Sarath
Multi-slice CT using STORESCP
Moderator: Moderator Team
-
- OFFIS DICOM Team
- Posts: 1446
- Joined: Tue, 2004-11-02, 17:22
- Location: Oldenburg, Germany
- Contact:
I don't know what "HERMES" is, but in DICOM CT slices are in most cases transmitted individually, because the traditional CT Image Storage SOP Class requires this. The new "Enhanced CT" SOP Class allows for a combination of multiple slices into one object, but is not supported yet by most modalities. StoreSCP, however, does support it (at least in the current version).
HiMarco Eichelberg wrote:I don't know what "HERMES" is, but in DICOM CT slices are in most cases transmitted individually, because the traditional CT Image Storage SOP Class requires this. The new "Enhanced CT" SOP Class allows for a combination of multiple slices into one object, but is not supported yet by most modalities. StoreSCP, however, does support it (at least in the current version).
HERMES is a processing workstation for Nuclear Medicine. It trasmits multi-CT to other DICOM servers without splitting. Any suggestions
Reagrds
Sarath
-
- OFFIS DICOM Team
- Posts: 1446
- Joined: Tue, 2004-11-02, 17:22
- Location: Oldenburg, Germany
- Contact: